Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • nao
    Junior Member
    • Aug 2015
    • 2

    #1

    Nextera XT tagmentation problem

    Hello all,
    I need some help with Nextera XT library prep from bacterial genome.
    DNA were extracted using QIAmp kit with QIAcube HT robot and measured the DNA by absorbance and fluorescence.
    I am following the Illumina protocol strictly and checking the library after PCR cleaning step to decide to move forward or start over the tagmentation & PCR.

    I had few samples to start over this time, but I am now stuck with tagmentation step for few days.
    I suspected inhibitor contamination or wrong DNA concentration, so cleaned the DNA with Zymo gDNA cleaning kit and eluted with water to avoid EDTA. DNA was measured by absorbance and fluorescence again to make sure the concentration. I confirmed the genomic DNA by running on QIAxcel.

    Brand new TD buffer and ATM buffer was used. After tagmentation (55 C, 5min), no tagmentation was going on confirming by QIAxcel.

    Does anyone had same experience? The DNA were all extracted under same condition at the same time, so why does this difference happens?

    In previous thread, it was recommended to clean with Mo-bio kit, so I will try that too.

    Any help or suggestions will be appreciated! I am now thinking to re-extract the DNA.
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Possible causes:
    1- Input DNA
    2- Reagent issues
    3- Reaction set up practices
    4- Thermocycler issues (can check the temperature with a probe or a thermometer)

    To narrow down the cause I would set up reactions using another DNA that has worked fine before or a clean commercial DNA with the same reagents that were used for failed reactions. It will be useful to post electropherogram of sample before and after failed tagmentation.

    Comment

    • nao
      Junior Member
      • Aug 2015
      • 2

      #3
      Hi nucacidhunter,

      Thank you for your suggestions. I didn't think about the thermocycler, but definitely will check that or run with other machine. I will also try with other working DNA to see how it goes.

      I prep 32 different DNA at a time and there are always some samples that don't go well. Usually it was some pipetting errors like that so when I repeat it from beginning, the final libraries look fine. This time I can't even pass the tagmentation step, which is very strange.

      I don't have electropherogram now, but will post once I save them.

      Comment

      • adam.geber
        Member
        • May 2014
        • 40

        #4
        Nextera XT requires input DNA of fairly low concentration (0.2ng/uL). What platforms are you using to measure your samples' concentration? Are you using QIAxcel readings to make your dilutions before tagmentation?

        As nucacidhunter mentioned, an electropherogram will be quite helpful. When you say that there is no tagmentation occurring do you mean that the traces from before and after look identical?

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
          by SEQadmin2




          NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
          10-30-2026, 06:46 AM
        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-28-2026, 09:57 AM
        0 responses
        36 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-28-2026, 09:51 AM
        0 responses
        22 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        72 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        48 views
        0 reactions
        Last Post SEQadmin2  
        Working...