  • Vgamma4
    Junior Member
    • May 2016
    • 2

    ATAC-seq small / overdigested fragments?

    Hi all,

    First, I hope that the ATAC-Family will grow further to share their experiences here.
    My experience with ATAC-sequence is kind of confuse. At the beginning I FACS sorted cells (20k - 50k) and followed the original Greenleaf paper. The fragments measured by BA were very small (one peak at 180bp). See Pic1

    Then I simplified the process by just taking freshly isolated cells from lymph nodes, tried different cell numbers from (10k - 100k), with and without lysis (as decried in other threads. Suddenly I got this typically nucleosome pattern and run those samples at the HiSeq (see Pic2). The results were promising. mtDNA reads were without lysis at 10% and with lysis around 40%. 25k cells seemed also optimal for sequencing.

    Then I went back with my optimised setting to the FACS sorting cells and got again only small fragments peak at 180bp. The sort by itself looked good.

    Could it be that sorting cells influences the quality of the libraries meaning that the cells are stressed or maybe already dying. Any experience with small fragments? Could it be that the DNA is overdigested?

    I am happy for any suggestions. Thanks
    Attached Files
  • mkel93
    Junior Member
    • Aug 2017
    • 1

    #2
    Hello,
    Did you ever optimise your protocol for sorted cells? I am having the exact same problem trying to do ATAC on 50K sorted haematopoietic cells (KSL).
    Thanks

    Comment

    • Page
      Junior Member
      • Jun 2018
      • 1

      #3
      Hey guys,

      Have you already figured out the possible reasons? I also just get short fragments (around 200 bp, 400 bp), however, without any long fragments. At beginning, we also suspected that the DNA is overdigested and we did series of tests within gradient dilution of Tn5. However, it doesn't help. We still didn't get the pattern and we just have a 200 bp peak.
      Any ideas? Do you think decreasing incubation time may help?

      Comment

      • xchrom
        Junior Member
        • Sep 2015
        • 1

        #4
        Was this ever resolved? I’m having a similar issue with primary cells. Only finding ~200bp fragments and no nucleosomal banding pattern. Any thoughts would be appreciated!

        Comment

        • OlgaMikh
          Junior Member
          • Feb 2012
          • 9

          #5
          Hello, has any of you solved this issue? Would greatly appreciate your response. I am seeing the same with FACS sorted cardiomyocyte nuclei.

          Comment

          • Meyana
            Member
            • Sep 2017
            • 40

            #6
            I have seen this on FACS'ed neuronal nuclei. Switching to the Omni-ATAC protocol (and handling the procedure a few times) solved everything and I now reliably get good looking BA tracks.
            When removing buffer after sorting, take care to remove the whole buffer volume. Any dilution of the tagmentation reaction by residual sorting buffer clearly affected my early efforts.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
              by SEQadmin2



              Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
              ...
              07-09-2026, 11:10 AM
            • SEQadmin2
              Cancer Drug Resistance: The Lingering Barrier to Rising Survival
              by SEQadmin2



              Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

              There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
              07-08-2026, 05:17 AM
            • GATTACAT
              Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
              by GATTACAT
              Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
              07-01-2026, 11:43 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 07-13-2026, 10:26 AM
            0 responses
            28 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-09-2026, 10:04 AM
            0 responses
            37 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-08-2026, 10:08 AM
            0 responses
            25 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-07-2026, 11:05 AM
            0 responses
            35 views
            0 reactions
            Last Post SEQadmin2  
            Working...