  • smsestes
    Junior Member
    • Feb 2020
    • 2

    bioanalyzer and qubit disagreement

    We prepare clinical samples with a formal SOP and do not regularly see irregularities. Today we have a capture pool of samples that quanted at 24.0ng/uL on qubit. To determine fragment size we also run sample pool in duplicate on bioanalyzer DNA1000. Duplicate 1, there is noise around 40bp (<1ng/uL) and no peak in the expected range (300-400). Duplicate 2, there is noise around 40bp and a wide peak between 700 and 1500 (upper marker). We reran the sample pool (4 replicates) on the bioanalyzer with only the noise around 40bp.

    Qubit concentration = normal
    bioanalyzer fragment size = missing?

    Any ideas what is going on? Where is the missing peak?
    Last edited by smsestes; 02-05-2020, 02:47 PM. Reason: clarity
  • JakobHedegaard
    Member
    • Mar 2008
    • 62

    #2
    Perhaps you didn't load enough on the Bioanalyzer to get a signal from a smear using a D1000 chip? Try to run on HS DNA chip.

    Comment

    • Noveline
      Junior Member
      • Jan 2020
      • 4

      #3
      DNA 1000 Kit is for 5-50 ng/ul of DNA
      Hs DNA Kit: "NGS sheared DNA or libraries: For accurate determination of DNA concentration, the total DNA in the samples must be between 100pg/μL to 10 ng/μL."

      I think that in HS dna chip will be overloaded...

      Check if everything is ok with gel dye mix in DNA 1000 Kit.

      Comment

      • smsestes
        Junior Member
        • Feb 2020
        • 2

        #4
        Originally posted by Noveline View Post
        DNA 1000 Kit is for 5-50 ng/ul of DNA
        Hs DNA Kit: "NGS sheared DNA or libraries: For accurate determination of DNA concentration, the total DNA in the samples must be between 100pg/μL to 10 ng/μL."

        I think that in HS dna chip will be overloaded...

        Check if everything is ok with gel dye mix in DNA 1000 Kit.
        Hi Noveline,

        When we reran the sample pool, we included a previous sample pool which ran normally and as expected. Thus it is unlikely to be the BioA or its reagents.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by SEQadmin2


          I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

          Here are nine questions we think about, in roughly the order they matter, before...
          06-18-2026, 07:11 AM
        • SEQadmin2
          From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
          by SEQadmin2


          Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


          The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
          ...
          06-02-2026, 10:05 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 06-17-2026, 06:09 AM
        0 responses
        34 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-09-2026, 11:58 AM
        0 responses
        99 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-05-2026, 10:09 AM
        0 responses
        120 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-04-2026, 08:59 AM
        0 responses
        112 views
        0 reactions
        Last Post SEQadmin2  
        Working...