    Index read problem

    Hi everyone, I'm a new user of Hiseq, this is a first time that I'm running a multiplex run and I have some concerns about Index read. Until the index read, the quality values were very high but after they dropped dramatically (see attached screenshot). Does it mean that I will not be able to demultiplex my samples?
    The multiplexing was done for one lane only (adapters 2, 4, 5, 6 and 12) and for all other lanes only one sample per lane was used. Can this be an issue? As far as I can conclude from this topic: http://seqanswers.com/forums/showthr...quality&page=2 this can also affect the quality of Read 2?
    For library preparation I used TruSeq DNA Sample Prep Kit, if this matters.
    Attached Files
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    You are probably fine. For the single index lanes, the reads that fail to de-multiplex will end up in the "unassigned" folder -- you can recover them from there.

    The focus carrying over to read 3 issue was solved with a firmware upgrade some months ago. Basically the instrument will always reset its focus at the beginning of read 3.

    --
    Phillip

    Comment

    • MLog
      Member
      • Jan 2010
      • 36

      #3
      Thanks, Phillip! Indeed, I checked it today and the Read3 seems to progress normally, with % of Q30 bases of about 95.

      And what about the multiplexed lane? Is it possible to retrieve some data from it? I assume that if index read quality is poor (only about 40% of Q30 bases) then there will be a lot of errors in the index sequences. CASAVA allows either no mismatches or 1 mismatch. What if we run some custom demultiplexing script, setting acceptable mismatch number to 2? Will it help or it will rather mix up the reads from different samples?
      Last edited by MLog; 11-28-2011, 02:38 AM.

      Comment

      • HESmith
        Senior Member
        • Oct 2009
        • 512

        #4
        Check the %Q30 for the lane that you multiplexed, rather than all lanes (as shown in attached figure). I suspect that it's fine.

        Comment

        • westerman
          Rick Westerman
          • Jun 2008
          • 1104

          #5
          [As HESmith implied in a post made while I was composing mine] Maybe you can send up a snapshot of the multiplex lane only? The snapshot you sent was all lanes thus is hard to tell what the quality of the index is like in the multiplex lane.

          Your barcodes are a good mix. None of them are very close to each other; i.e., have similar bases. On the other hand with a mismatch of 2 you can get into problems with choosing which barcode to choose. For example a read that has 'CGATCT' could, with 2 mismatches, go to either barcode #2 or #4.

          A better bet, if possible, is to just not use all 6 bases. It is possible in CASAVA to mask out which bases to use in the barcode. I've done this when the quality in some of the index bases was poor but the quality was high in the other bases. It is possible to even go down to one base for demulitplexing if said base is very high quality -- obviously with no mismatches allowed. Desperate times require desperate measures!
          Last edited by westerman; 11-28-2011, 08:11 AM. Reason: Added comment about HESmith

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            By the way, you can right click on most SAV graphs and the context menu will offer the choice "copy to clipboard". Might be easier than taking a picture of the screen...

            --
            Phillip

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            28 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            21 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-15-2026, 10:52 AM
            0 responses
            38 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            44 views
            0 reactions
            Last Post SEQadmin2  
            Working...