    How to mix single index and dual index TruSeq libraries

    In principle I see no issues with mixing dual and single index TruSeq libraries in the same run. The result will be that for the i7 read, the normal single-index sequence will be collected along with an extra 2 bases. For indexes 1-12, these last 2 bases will always be "AT".

    I think their i7 sequences (as would be entered in the sample sheet would be as follows) :

    >TruSeq_Single_Index_1
    ATCACGAT
    >TruSeq_Single_Index_2
    CGATGTAT
    >TruSeq_Single_Index_3
    TTAGGCAT
    >TruSeq_Single_Index_4
    TGACCAAT
    >TruSeq_Single_Index_5
    ACAGTGAT
    >TruSeq_Single_Index_6
    GCCAATAT
    >TruSeq_Single_Index_7
    CAGATCAT
    >TruSeq_Single_Index_8
    ACTTGAAT
    >TruSeq_Single_Index_9
    GATCAGAT
    >TruSeq_Single_Index_10
    TAGCTTAT
    >TruSeq_Single_Index_11
    GGCTACAT
    >TruSeq_Single_Index_12
    CTTGTAAT
    >TruSeq_Single_Index_13
    AGTCAACA
    >TruSeq_Single_Index_14
    AGTTCCGT
    >TruSeq_Single_Index_15
    ATGTCAGA
    >TruSeq_Single_Index_16
    CCGTCCCG
    >TruSeq_Single_Index_18
    GTCCGCAC
    >TruSeq_Single_Index_19
    GTGAAACG
    >TruSeq_Single_Index_20
    GTGGCCTT
    >TruSeq_Single_Index_21
    GTTTCGGA
    >TruSeq_Single_Index_22
    CGTACGTA
    >TruSeq_Single_Index_23
    GAGTGGAT
    >TruSeq_Single_Index_25
    ACTGATAT
    >TruSeq_Single_Index_27
    ATTCCTTT

    Then the i5 read will always be "TCTTTCCC" -- as that is the segment of the left adapter that would be addressed by the dual index primer for i5.


    I brought this up in a previous thread. Therein, a poster, "Vinz", specified exactly what to do in a manner so succinct I had no idea how to interpret it.

    Note: this post contains information partially derived from the Illumina Customer Sequence Letter. Illumina specifies in that letter that they give limited permission to "distribute the sequence information outside your institution or to publish the sequence information in presentations, manuscripts, or publications authored by you, as long as it is accompanied by the following copyright notice:"

    Oligonucleotide sequences © 2007-2012 Illumina, Inc. All rights reserved.

    --
    Phillip
  • kcchan
    Senior Member
    • Jul 2012
    • 186

    #2
    In theory you should be able to set index 2 of the single indexed reads as NNNNNNNN and it should demultiplex just fine. If each of your dual indexed samples have a unique index 1 you can even skip the second index read entirely.

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      Philip,

      I assume you are asking if this will work if the analysis is being done by MiSeq/BaseSpace. We never analyze the data on our MiSeq (or using BaseSpace for that matter) so I will only speculate about using CASAVA to analyze the data after the run.

      Based on what you have said above it should be possible to de-multiplex this run since the tag pairs should all be unique (but pair would be meaningless for those samples which are single indexed).

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        At this point I am pretty sure it will work and it is obvious how to create a sample sheet that will cause the dual indexed and single indexed libraries to demultiplex.

        Illumina tech support initially claimed it was not possible, then when I insisted that it should be, backed off to saying we would have to re-run MSR a second time after hacking runinfo.xml to pull it off. But even that is not necessary and the fix is pretty trivial.

        --
        Phillip

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #5
          Originally posted by pmiguel View Post
          At this point I am pretty sure it will work and it is obvious how to create a sample sheet that will cause the dual indexed and single indexed libraries to demultiplex.

          Illumina tech support initially claimed it was not possible, then when I insisted that it should be, backed off to saying we would have to re-run MSR a second time after hacking runinfo.xml to pull it off. But even that is not necessary and the fix is pretty trivial.

          --
          Phillip
          Do report back and confirm if it does work. Would save a new thread from being created for this question.

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by GenoMax View Post
            Do report back and confirm if it does work. Would save a new thread from being created for this question.
            Yes, it works fine.

            The run itself was less than stellar, but Rick was able to use CASAVA to demultiplex both the single index and the dual index clusters in a single pass.

            --
            Phillip

            Comment

            • GenoMax
              Senior Member
              • Feb 2008
              • 7142

              #7
              Originally posted by pmiguel View Post
              Yes, it works fine.

              The run itself was less than stellar, but Rick was able to use CASAVA to demultiplex both the single index and the dual index clusters in a single pass.

              --
              Phillip
              Was this done by CASAVA analysis after the run or on the MiSeq itself?

              Comment

              • pmiguel
                Senior Member
                • Aug 2008
                • 2328

                #8
                Originally posted by kcchan View Post
                In theory you should be able to set index 2 of the single indexed reads as NNNNNNNN and it should demultiplex just fine. If each of your dual indexed samples have a unique index 1 you can even skip the second index read entirely.
                Or "TCTTTCCC" -- that is the sequence that gets read if you do and index2 (i5) read on a TruSeq single index library. (Well, if you are doing PE reads. For SR flowcells, I guess you would get the reverse complement of that because the second index has to be read on the same strand as the first index.)

                --
                Phillip

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by GenoMax View Post
                  Was this done by CASAVA analysis after the run or on the MiSeq itself?
                  It was done with CASAVA. But I think it would have worked fine on the MiSeq as well if the correct sample sheet had been provided it.

                  These were amplicon samples generated based on a design I provided the investigator. I inadvertently designed them using the reverse complement of the desired "A" series dual index i7 sequences. Hence, nothing demultiplexed on the MiSeq.

                  I think it might be possible to force the MiSeq to repeat the entire analysis using a new sample sheet. But I have not done this before. Have you?

                  --
                  Phillip

                  Comment

                  • microgirl123
                    Senior Member
                    • Jun 2012
                    • 199

                    #10
                    We have a pretty new MiSeq and I spend a lot of time changing the sample sheet and telling MSR to rerun the analysis!

                    Comment

                    • GenoMax
                      Senior Member
                      • Feb 2008
                      • 7142

                      #11
                      Originally posted by pmiguel View Post

                      I think it might be possible to force the MiSeq to repeat the entire analysis using a new sample sheet. But I have not done this before. Have you?

                      --
                      Phillip
                      I have done it only once as an exercise on a PC where I installed MiSeq reporter standalone for testing. I am reasonably certain that you should be able to repeat the analysis on the MiSeq. I suggest making a backup copy of the data folder before trying it

                      Comment

                      • pmiguel
                        Senior Member
                        • Aug 2008
                        • 2328

                        #12
                        Originally posted by microgirl123 View Post
                        We have a pretty new MiSeq and I spend a lot of time changing the sample sheet and telling MSR to rerun the analysis!
                        Want to give a quick run down of how you do that? Would be appreciated...

                        --
                        Phillip

                        Comment

                        • kcchan
                          Senior Member
                          • Jul 2012
                          • 186

                          #13
                          It's quite easy to edit a sample sheet in MSR. Fire up Internet Explorer on the MiSeq and go to localhost:8042/ In the Analyses tab, select the run you want to edit. Then go to the Sample Sheet tab and edit the file like you would in Excel. Here you can change the workflow, sample names, adapter sequences, and index sequences. When you're done just click "Save and Requeue" and let the MiSeq crunch the data!

                          Comment

                          • pmiguel
                            Senior Member
                            • Aug 2008
                            • 2328

                            #14
                            Originally posted by kcchan View Post
                            It's quite easy to edit a sample sheet in MSR. Fire up Internet Explorer on the MiSeq and go to localhost:8042/ In the Analyses tab, select the run you want to edit. Then go to the Sample Sheet tab and edit the file like you would in Excel. Here you can change the workflow, sample names, adapter sequences, and index sequences. When you're done just click "Save and Requeue" and let the MiSeq crunch the data!
                            Yes, that does sound easy. However when I browse to that URL, the Analyses tab is is empty.

                            I checked and the directory it is pointing to does appear to be the correct one. Any ideas?

                            --
                            Phillip

                            Comment

                            • GenoMax
                              Senior Member
                              • Feb 2008
                              • 7142

                              #15
                              Originally posted by pmiguel View Post
                              Yes, that does sound easy. However when I browse to that URL, the Analyses tab is is empty.

                              I checked and the directory it is pointing to does appear to be the correct one. Any ideas?

                              --
                              Phillip
                              Did you try looking at a previous run (since you had said that the most recent run was analyzed off-line by CASAVA)?

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                10-30-2026, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              45 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              28 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              76 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              49 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...