    Hi, All:

    New to the forum here. I was running Miseq PE with 1 indexing (read1 14 cycles, index 6, read2 50 cycles), and the run finished without issue. However, during Miseq reporter analysis, only the 50 cycle reads were demultiplexed and fastq generated, while no 14 cycle reads fastq was generated.

    This is what the sample sheet looks like:

    Workflow,GenerateFASTQ
    Application,FASTQ Only
    Assay,TruSeq LT
    Description,
    Chemistry,Default

    [Reads]
    14
    50

    [Settings]

    [Data]
    Sample_ID,Sample_Name,Sample_Plate,Sample_Well,I7_Index_ID,index,Sample_Project,Description
    sample1,1,sample plate,A01,A006,GCCAAT
    sample2,2,sample plate,A02,A012,CTTGTA

    Has anyone experienced similar issue? Do I need to use CASAVA to do my own demultiplexing? Thanks a lot!
  • kcchan
    Senior Member
    • Jul 2012
    • 186

    #2
    It may have something to do with RTA not getting enough data to perform basecalling. The first 12 bases are used to calculate the color matrix and phasing/prephasing values, and 25 are needed for quality scoring.

    Just curious, may I ask why this run is designed to be in this asymmetric manner?
    Last edited by kcchan; 03-10-2013, 07:39 PM.

    Comment

    • sunleaf
      Junior Member
      • Mar 2011
      • 3

      #3
      Thanks for your answer. That's what I suspected too. We are developing our own assay and 14 cycle is all we needed for that end. I'll contact Illumina to see if I can salvage the data and run 25 cycles in the future.

      Comment

      • kcchan
        Senior Member
        • Jul 2012
        • 186

        #4
        You can check the error logs in MSR and see what's going on. I would guess that BCL files were never generated from read 1 and all you had were CIF intensity files. You might be able to use offline basecaller to get the basecalls out of it, but I've never tried it in such a situation. Illumina tech support may be your best bet in this case. Good luck and let us know how things turn out!

        Comment

        • sunleaf
          Junior Member
          • Mar 2011
          • 3

          #5
          Just talked with a very helpful Illumina tech support. Yes, we can use CASAVA to demultiplex the other end with --usebasemask option and designated 14 bases, and the sample sheet should be made for Hiseq instead of Miseq. That's just too much trouble to go through for our repeated runs. I think it would just be easier to run read1 for 25 cycles.

          Comment

          • kmcarr
            Senior Member
            • May 2008
            • 1181

            #6
            Originally posted by sunleaf View Post
            Just talked with a very helpful Illumina tech support. Yes, we can use CASAVA to demultiplex the other end with --usebasemask option and designated 14 bases, and the sample sheet should be made for Hiseq instead of Miseq. That's just too much trouble to go through for our repeated runs. I think it would just be easier to run read1 for 25 cycles.
            If you are using a 50 cycle kit make sure that the total number of cycles doesn't exceed the capacity of the kit. Officially a 50 cycle kit has sufficient reagents for 76 cycles; we have safely done 78 cycles. If you increase read 1 to 25 cycles without changing the other settings you cycle total will be 81 (25|6|50).

            Comment

            • kcchan
              Senior Member
              • Jul 2012
              • 186

              #7
              Good point kmcarr. It's never good to run out of reagents since it can suck air into the lines. I believe the scan mix is usually the first to run out.

              It may be a good idea to re-think the library design a bit. You can squeeze the indices inline at the start of read 1 to lengthen the read, but unfortunately I don't think the MiSeq has an easy way to demultiplex inline barcodes, so you're stuck with CASAVA if you want to go that route.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                by SEQadmin2




                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                Today, 06:46 AM
              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                10-14-2026, 09:55 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 10-28-2026, 09:57 AM
              0 responses
              22 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-28-2026, 09:51 AM
              0 responses
              14 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 11:13 AM
              0 responses
              67 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 10:48 AM
              0 responses
              47 views
              0 reactions
              Last Post SEQadmin2  
              Working...