    Library failure - any help appreciated

    Hi all

    I am having trouble getting a MiSeq run to work and was hoping someone might be able to help

    I have swab samples from surfaces in a hospital, i have DNA extracted using a bead beat and Qiagen column. Then I have pooled the extracted DNA from bed spaces and PCR's using 16S primers and custom barcodes for my library prep.

    I have qPCR'd almost all of the samples individually before pool with a Staph generic assay and there are a few with not much in and some that may have been inhibited but I would think that would be diluted out with the pooling.

    So, I have PCR'd to add barcodes, run a size select gel to purify - seen bands on all lanes. I have Qubited, adjusted to 4nM, pooked 96 samples, qubited again and then used this as my library. Denatured and prepared as per Illumina protocol and the samples are not working.

    Does anyone have any ideas? thanks
  • Heisman
    Senior Member
    • Dec 2010
    • 534

    #2
    What do you mean by not working? How do you know they failed?

    Comment

    • firefly2280
      Member
      • Oct 2012
      • 28

      #3
      Originally posted by Heisman View Post
      What do you mean by not working? How do you know they failed?
      The Phi X has clustered but the run then fails, so no reads at all from my samples.

      Comment

      • kwaraska
        Senior Member
        • Nov 2008
        • 131

        #4
        I would confirm the sequences you put on with the barcodes.

        "Old style"-non Truseq single end adapters don't work on the MiSeq although they do work on the HiSeq.

        Comment

        • firefly2280
          Member
          • Oct 2012
          • 28

          #5
          Originally posted by kwaraska View Post
          I would confirm the sequences you put on with the barcodes.

          "Old style"-non Truseq single end adapters don't work on the MiSeq although they do work on the HiSeq.
          Thanks, they are the exact same ones as being used by other MiSeq users in my lab and they are getting good results

          Comment

          • GW_OK
            Senior Member
            • Sep 2009
            • 411

            #6
            Can you post SAV specs? Cluster density, %PF, base intensity plots....

            Comment

            • firefly2280
              Member
              • Oct 2012
              • 28

              #7
              Originally posted by GW_OK View Post
              Can you post SAV specs? Cluster density, %PF, base intensity plots....
              Hi, sadly not as the runs failed. the samples failed to cluster, nothing worked. apart from the Phi X.

              Comment

              • GenoMax
                Senior Member
                • Feb 2008
                • 7142

                #8
                Originally posted by firefly2280 View Post
                Hi, sadly not as the runs failed. the samples failed to cluster, nothing worked. apart from the Phi X.
                If phiX worked (I assume that means you got the sequence) then you should have the graphs/stats that GW_OK asked you to post. They will only show phiX data.

                I have come across only one instance of a sample where it would not work on a HiSeq after multiple runs but did work on a MiSeq. Perhaps if you have access to a HiSeq you could try running your samples there (if the libraries are compatible).
                Last edited by GenoMax; 07-22-2013, 06:55 AM.

                Comment

                • firefly2280
                  Member
                  • Oct 2012
                  • 28

                  #9
                  Originally posted by GenoMax View Post
                  If phiX worked (I assume that means you got the sequence) then you should have the graphs/stats that GW_OK asked you to post. They will only show phiX data.

                  I have come across only one instance of a sample where it would not work on a HiSeq after multiple runs but did work on a MiSeq. Perhaps if you have access to a HiSeq you could try running your samples there (if the libraries are compatible).
                  ok, thanks will do later. I dont have anything with me. Im really new to this (you can probably tell)!

                  Comment

                  • Number6
                    Member
                    • Feb 2009
                    • 19

                    #10
                    If you have a custom primer for your sample, the MiSeq requires a higher Tm than primers on the HiSeq (I believe this is discussed in other threads in this forum). We had a sample that only sequenced on the MiSeq when we used a longer sequencing primer, and failed with a shorter version of the same primer. Might be worth checking the Tm (if applicable).

                    Comment

                    • firefly2280
                      Member
                      • Oct 2012
                      • 28

                      #11
                      Originally posted by Number6 View Post
                      If you have a custom primer for your sample, the MiSeq requires a higher Tm than primers on the HiSeq (I believe this is discussed in other threads in this forum). We had a sample that only sequenced on the MiSeq when we used a longer sequencing primer, and failed with a shorter version of the same primer. Might be worth checking the Tm (if applicable).
                      Many thanks Number6. The primers are all being used successfully by a colleague, so we dont think they are the issue. Thanks though

                      Comment

                      Latest Articles

                      Collapse

                      • SEQadmin2
                        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                        by SEQadmin2



                        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                        ...
                        10-14-2026, 09:55 AM
                      • SEQadmin2
                        New Genomics Technologies Take Aim at Long-Standing Limits
                        by SEQadmin2


                        Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                        We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                        ...
                        09-28-2026, 10:25 AM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, 10-15-2026, 10:52 AM
                      0 responses
                      14 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-14-2026, 05:39 AM
                      0 responses
                      21 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-13-2026, 09:56 AM
                      0 responses
                      20 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-09-2026, 11:14 AM
                      0 responses
                      27 views
                      0 reactions
                      Last Post SEQadmin2  
                      Working...