Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • haroldnunez
    Junior Member
    • Jan 2014
    • 5

    #1

    MiSeq run with low Q30 and high cluster PF

    Hi all,

    Could anybody give me an idea or though on the following?

    I made a run with the MiSeq using the TrueSeq LT kit and 151 cycles. It was an amplicon library with 3 different DNA, everyone with his own index. The run has very low Q30, around 6%, but has Clusters PF (passing filter) of 85%. The number of clusters is around 1100 K/mm2. Moreover, no reads were identified.

    Could have been the PhiX? it was set at 15% but not freshly made, only two weeks old (protocol recommendation is 3 weeks max).

    Anyway, any help will be greatly appreciated. Bye.
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    You should contact Illumina tech support. They should be able to do a remote login and look at this run. That may be the fastest way to diagnose what is going on with this run.

    That said: What do you mean by "no reads were identified (no sequence or did the samples not demultiplex)" when you had 85% clusters that passed filter (leaving the Q30 aside for now since this was an amplicon run).

    Comment

    • haroldnunez
      Junior Member
      • Jan 2014
      • 5

      #3
      Originally posted by GenoMax View Post
      That said: What do you mean by "no reads were identified (no sequence or did the samples not demultiplex)" when you had 85% clusters that passed filter (leaving the Q30 aside for now since this was an amplicon run).
      Thanks for your advice. About your comment, it should have said no Index were identified. I have 18,096,508 reads but none is assigned to an index.

      Anyone has experienced something like this before?

      Comment

      • GenoMax
        Senior Member
        • Feb 2008
        • 7142

        #4
        If you are sure that the index read has worked and for some reason the demultiplexing did not then you can:

        a) Try to re-run the analysis with MiSeq reporter or if that does not work
        b) try doing the basecalling/demultiplexing using bcltofastq on a separate unix server (for ref: http://seqanswers.com/forums/showthread.php?t=39153)

        In any case first see what Tech Support says (I see that you are in Chile but hopefully there is a way to contact the tech support either local or the one in US).

        Comment

        • haroldnunez
          Junior Member
          • Jan 2014
          • 5

          #5
          Originally posted by GenoMax View Post
          If you are sure that the index read has worked and for some reason the demultiplexing did not then you can:

          a) Try to re-run the analysis with MiSeq reporter or if that does not work
          b) try doing the basecalling/demultiplexing using bcltofastq on a separate unix server (for ref: http://seqanswers.com/forums/showthread.php?t=39153)

          In any case first see what Tech Support says (I see that you are in Chile but hopefully there is a way to contact the tech support either local or the one in US).
          I will try your suggestion, however I'm worried about the quality score. It could improve if I do the basecalling/demultiplexing with bcltofastq? or I just will recover the index?

          Comment

          • GenoMax
            Senior Member
            • Feb 2008
            • 7142

            #6
            Was this a reagent V2 or V3 run? 1100 clusters/mm^2 may be bad news for the tag read if these are V2 reagents. You may have to repeat the run in that case. Tech support can determine if the run was overloaded by looking at the thumbnails.

            Comment

            • haroldnunez
              Junior Member
              • Jan 2014
              • 5

              #7
              Originally posted by GenoMax View Post
              Was this a reagent V2 or V3 run? 1100 clusters/mm^2 may be bad news for the tag read if these are V2 reagents. You may have to repeat the run in that case. Tech support can determine if the run was overloaded by looking at the thumbnails.
              Bad news indeed, it was a V2 run. Some thumbnails.

              Click image for larger version

Name:	s_1_1111_a.jpg
Views:	1
Size:	76.7 KB
ID:	304403

              Click image for larger version

Name:	s_1_1111_c.jpg
Views:	1
Size:	70.3 KB
ID:	304404

              Click image for larger version

Name:	s_1_1111_g.jpg
Views:	1
Size:	78.1 KB
ID:	304405

              Click image for larger version

Name:	s_1_1111_t.jpg
Views:	1
Size:	73.3 KB
ID:	304406

              Comment

              • GenoMax
                Senior Member
                • Feb 2008
                • 7142

                #8
                You can push cluster numbers beyond the recommended limits (http://www.illumina.com/systems/mise...fications.ilmn) for normal sequencing but in case of multiplexing one runs the risk on losing the run (like in this case, even though actual sequence reads are probably acceptable).

                If these are amplicons you should stay within (lower end of) the recommended limits (800-900 clusters/mm^2 for V2).
                Last edited by GenoMax; 01-08-2014, 12:36 PM.

                Comment

                • ECO
                  --Site Admin--
                  • Oct 2007
                  • 1360

                  #9
                  Look at the base diversity of read 1.

                  Comment

                  • haroldnunez
                    Junior Member
                    • Jan 2014
                    • 5

                    #10
                    Originally posted by ECO View Post
                    Look at the base diversity of read 1.
                    Is way off to A and T. I should repeat it, however it is wise tu use the same library?

                    Could PhiX be the problem as stated before?

                    Comment

                    • GenoMax
                      Senior Member
                      • Feb 2008
                      • 7142

                      #11
                      Can you post a FastQC plot of the distribution?

                      Irrespective of not being able to demultiplex the data does the sequence you have look to be the right mix of what you expect? If it is not then you should think twice about resequencing the same library.

                      Comment

                      • Shimul
                        Junior Member
                        • Jan 2017
                        • 8

                        #12
                        Hi There,

                        I am running my bacterial sample using MiSeq V3. After 70 of 600 cycles, it showed cluster density is 1092K/mm, cluster passing filter 91.4% and the >=Q30 is 96%. Do you think that the run is going well?

                        Comment

                        • GenoMax
                          Senior Member
                          • Feb 2008
                          • 7142

                          #13
                          Originally posted by Shimul View Post
                          Hi There,

                          I am running my bacterial sample using MiSeq V3. After 70 of 600 cycles, it showed cluster density is 1092K/mm, cluster passing filter 91.4% and the >=Q30 is 96%. Do you think that the run is going well?
                          So far yes.

                          Comment

                          Latest Articles

                          Collapse

                          • SEQadmin2
                            16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                            by SEQadmin2




                            NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                            10-30-2026, 06:46 AM
                          • SEQadmin2
                            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                            by SEQadmin2



                            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                            ...
                            10-14-2026, 09:55 AM

                          ad_right_rmr

                          Collapse

                          News

                          Collapse

                          Topics Statistics Last Post
                          Started by SEQadmin2, 10-28-2026, 09:57 AM
                          0 responses
                          40 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 10-28-2026, 09:51 AM
                          0 responses
                          23 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 10-22-2026, 11:13 AM
                          0 responses
                          73 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 10-22-2026, 10:48 AM
                          0 responses
                          48 views
                          0 reactions
                          Last Post SEQadmin2  
                          Working...