  • neko_33
    Junior Member
    • Dec 2016
    • 3

    #1

    Terrible overclustering -- 16S EMP protocol

    Hi SEQ-users,

    We've been using the following protocol for 16S V4 region amplicon sequencing on MiSeq, which I think is very familiar to many of you:
    DNA sequencing continues to decrease in cost with the Illumina HiSeq2000 generating up to 600 Gb of paired-end 100 base reads in a ten-day run. Here we present a protocol for community amplicon sequencing on the HiSeq2000 and MiSeq Illumina ...


    For several years, most members in our lab have been constantly successful in that method.
    Recently, however, cluster density is abnormally high -- about 1300~1500K/mm2(v2 kit) -- and sequencing quality very low. We tried over and over, reducing the library concentration from 15pM to 10pM, renewing sequencing primers, renewing phiX controls, but none of these worked.

    In additon, libraries prepared with other methods (e.g. 2-step PCR, NEB UltraDNA Library Prep Kit, bacterial RNA-seq, etc.) are still sequenced fairly well. This probably means that, our trouble is not due to the MiSeq instrument or our poor sample handling.

    Do any of you have such problems?
    Or, does anyone knows the solution?

    Thanks a lot in advance.
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Most likely reason for overclustering in this case would be inaccurate library quantification and high clustering of low diversity library will result in low quality. If reducing loading from 15 pM to 10 pM reduced cluster number then you might review the quantification.

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      We recently had a set of libraries (not 16S, but targeted loci, so very similar) that read as 4nM on the bioanalyzer but 20nM via qPCR! These were so discrepant that we repeated both assays at least twice with similar results each time.
      We used the qPCR numbers and ended up at ~700Kclusters/mm^2 on a v2 MiSeq cassette. A little low, but far preferable to overshooting the cluster density.
      neko_33 what method are you using to quantitate your library concentration? Also, if you are using phiX as your standards for qPCR, don't do that! The phiX standards are notorious for varying in concentration from batch to batch.

      --
      Phillip

      Comment

      • neko_33
        Junior Member
        • Dec 2016
        • 3

        #4
        quantification failure? -- maybe not

        Thank you for your kind replies.

        We usually quantify libraries with Qubit fluorometer, but at times with qPCR (KAPA Library Quantification Kits). Illumina Co. warns that Biolanalyzer quanitification is not reliable, according to my memory.

        Empirically, not so much difference between Qubit and qPCR. For example, when I last did MiSeq run, the concentration calculated based on Qubit result was 1.9±0.1nM, while 2.2nM on qPCR.

        As I wrote in the first message, I think our quantifications do not have thus fatal errors, but I will continue using both Qubit and qPCR, and compare these two results.

        Comment

        • thermophile
          Senior Member
          • Apr 2015
          • 243

          #5
          Is this the same machine? Each machine clusters differently (I'd guess due to fluctuations in temperature control). FWIW, I cluster 16S runs at 6pM. When I go up to even 8pM I end up loosing so many clusters at PF that the output is roughly the same.
          Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

          Comment

          • neko_33
            Junior Member
            • Dec 2016
            • 3

            #6
            Dear thermophile,

            Thank you very much for your suggestion. The machine we use is always the same.
            But I will check if the temperature is decently controled.

            In addition, it is surprising that even 8pM library is too high!

            Comment

            • thermophile
              Senior Member
              • Apr 2015
              • 243

              #7
              The EMP sequencing primers aren't the right TM (they're too low). I've had the opposite problem you've had in the past-regular libraries will sequence fine but I'll get very very few clusters on the 16s runs. This was on an older miseq and the temperature control module wasn't hitting and holding temps correctly. It was still within the error tolerance of illumina libraries but the emp libraries have a narrower tolerance.
              Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                by SEQadmin2




                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                Today, 06:46 AM
              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                10-14-2026, 09:55 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 10-28-2026, 09:57 AM
              0 responses
              24 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-28-2026, 09:51 AM
              0 responses
              15 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 11:13 AM
              0 responses
              68 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 10:48 AM
              0 responses
              48 views
              0 reactions
              Last Post SEQadmin2  
              Working...