    Reason for POOR per tile sequence quality

    We recently performed a sequencing run on the MiSeq sequencer using a V3 (600 cycle kit).

    With regards to the per tile quality plot (obtained when using the FastQC analysis software) for one of our samples, it seems as though there is a large area that is flagged in red for R2, indicating that those tiles had worse qualities.

    I was just wondering on possible reasons why this would happen and how it could be prevented in future runs?

    Thanks!
    Attached Files
  • Markiyan
    Senior Member
    • Sep 2010
    • 126

    #2
    Looks like there is a problem with imaging one of the flowcell surfaces...

    It looks like there is a problem with imaging one of the flowcell surfaces...

    Either focusing issue (Z axis lens mover in a need of some cleaning/re greasing/calibration) or lots of very small bubbles (if it is an upper surface) (but then they would affect the other surface with such width.

    Also check tile thumbnail images, and contact illumina if this instrument is on service contract.

    You can also try making movie from thumbnail jpg images using image magic and some scripting...

    Comment

    • GW_OK
      Senior Member
      • Sep 2009
      • 411

      #3
      2xxx tiles are on the bottom surface. I concur with the possibilities raised by Markiyan. Looks like it started somewhere around cycle 35. Definitely a call to tech support is in order.

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        The Factors Reshaping Genome Assembly
        by SEQadmin2



        When the Human Genome Project declared the human genome finished in 2003, portions of the sequence were still missing, including many of its most repetitive and structurally complex regions.1 Still, for the next two decades, researchers analyzed their sequencing data against the single linear reference that came out of this effort. Released in 2013 as GRCh38, that reference has powered countless discoveries, but it has gaps and cannot capture the genetic diversity of the world’s...
        11-24-2026, 07:04 AM
      • SEQadmin2
        16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
        by SEQadmin2





        NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification,...
        10-30-2026, 06:46 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, 11-23-2026, 11:11 AM
      0 responses
      39 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 11-17-2026, 06:27 AM
      0 responses
      47 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 11-12-2026, 07:59 AM
      0 responses
      68 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 11-05-2026, 09:25 AM
      0 responses
      68 views
      0 reactions
      Last Post SEQadmin2  
      Working...