Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • bsvega
    Member
    • Jul 2015
    • 10

    #1

    NEBNext adaptors clustering failed

    Hi all,
    I am having problems with the NEBnext Ultra kit for library preparation. I´ve been doing same amplicon libraries for three years with success, but since October none of my libraries seems to hybridize with the adapters of the flow cell (I got 0 clusters). It happened already with 6 different runs (Miseq v3, 600 cycles) and 240 different library preps.
    The bioanalyzer pattern was good (size OK, no adapters, shape OK), final pools were quantified by Qubit, Bionalyzer and qPCR (Qiaseq). None of the quantifications worked. PhiX was loaded at 10 and 15% equimolar to pooled libraries (12pM).
    NaOH and Miseq are working because other researchers are using them with no problems.
    Any idea of what can be the problem?. Neither Illumina, nor NEB, are helpful.
    Thanks
  • Markiyan
    Senior Member
    • Sep 2010
    • 126

    #2
    Check sequencing adapters.

    Try a new batch of the sequencing adapters.
    Maybe one of your sequencing adapters had degraded?
    Also check the thumbnail images from the run's tmp folder.

    If both adapters at the end of the insert are same P5 or P7, than the qPCR would work, but clustering wouldn't.

    Comment

    • bsvega
      Member
      • Jul 2015
      • 10

      #3
      Hi Markiyan, thanks for your reply.
      I already use two different lots of adapters (and reagents).
      There are no thumbnail images, the folder is empty.
      The Mise, althought it says Run completed, the runs stoped before completion (before 24h for a run of 600 cyles).

      Comment

      • Markiyan
        Senior Member
        • Sep 2010
        • 126

        #4
        A couple things to check:

        1. Did you look into
        d:\Illumina\MiSeqTemp\[your run ID]\Thumbnail_Images ?
        It is available only till the start of the next run.

        2. Did you use PhiX in 10-15%? - at least it should show the reads from the PhiX...

        3. PS: If troubleshooting - use 300 or 500 cycles nano kits (cheaper than 600 cycles V3).

        Comment

        • nucacidhunter
          Jafar Jabbari
          • Jan 2013
          • 1250

          #5
          Originally posted by bsvega View Post
          The bioanalyzer pattern was good (size OK, no adapters, shape OK), final pools were quantified by Qubit, Bionalyzer and qPCR (Qiaseq). None of the quantifications worked.

          if I understand correctly, individual libraries BA profile was fine but the library pool did not have any DNA according to Qubit, BA and qPCR?

          If library PCR amplification post adapter ligation was succssfull, adapters are unlikely to be the cause.

          You have mentioned that you are preparing amplicon libraries. It could be due to polymerase if you now using different one. More information on your library prep workflow would give clues for troubleshooting.

          Comment

          • bsvega
            Member
            • Jul 2015
            • 10

            #6
            Yes, I thought about v3-600 cycles problem, so I tried with a Miseq nano v2. Same problem, no clustering.

            Idividual libraries and pool bionalyzer profiles are good. Quantification of individual libraries is done with qubit, quantification of pooled libraries is done with qubit, bioanalyzer and qPCR.

            My libraries are IGH genes. I do a PCR amplification from gDNA using an international estandarized protocol (Biomed-2), so it use to work with no problems. I clean-up the PCR products with Ampure XP beads and quantitate the ammount of product used for the library normally between 100 and 500 ng. I follow the NEBNext protocol with no changes.

            The bionalyzer profile between PCR products and individual libraries show a difference of around 120-130 bp. And because qPCR worked I think the libraries are OK, but I do not get any clusters.
            Do not know what else to try It used to work perfectly before.

            Comment

            • Markiyan
              Senior Member
              • Sep 2010
              • 126

              #7
              PhiX% ?

              1. Do you use any custom sequencing primers (esp R1)? If yes - more detail where and for which reads.

              2. HOW MUCH PhiX % did you load!?? - I hope that at least 10%?

              PS: there are some GC stretches in those reads - so please also check the thermal cycling performance of the instrument during the cluster generation.

              Comment

              • GW_OK
                Senior Member
                • Sep 2009
                • 411

                #8
                If you have a shift in peak size of ~125bp between your initial amplicon and the final library you make from it then it should have the appropriate adapters ligated to each end.

                If your qPCR of these libraries is "not working" you should stop immediately and investigate. I think (at least for the Kapa kit I use) the qPCR primers are based off of the p5 and p7 sites and if you cannot qPCR it then you have no hope of generating clusters.

                If you are loading 10% PhiX (as mentioned in the first post) it is very odd that you are seeing a failure in clustering. Two things that might cause this are accidentally telling the Miseq to use a custom primer when you don't want it to (leading to lack of priming and nothing detected) or overclustering (as too many clusters may look like no clusters if focus fails).

                Comment

                • nucacidhunter
                  Jafar Jabbari
                  • Jan 2013
                  • 1250

                  #9
                  Good points have been raised in posts above. A run with 40% PhiX spike in should clear whether your library has any issues or if fail is due to preparation and sequencing set up.

                  Comment

                  • bsvega
                    Member
                    • Jul 2015
                    • 10

                    #10
                    Thank you for your input. I wil try to answer everything.

                    - I don´t use custom primers because NebNext does not need them.
                    - qPCR, bionalyzer and Qubit work for individual libraries and also for pooled libraries. qPCR quantification used to give a slighty higher concentration (I guess because of heteroduplex from these are low diversity libraries). I run the same pool twice according to qPCR and Qubit quantifications with same results.
                    - I used to load 5% Phix with these libraries and worked fine. Since the failed runs I am loading 10 and 15%. I dont get anyclusters, neither from PhiX. As GW_OK said are accidentally telling the Miseq to use a custom primer when you don't want it to (leading to lack of priming and nothing detected). How can I check this?. The samplesheet does not specify to use custom primers.

                    " Markiyan : PS: there are some GC stretches in those reads - so please also check the thermal cycling performance of the instrument during the cluster generation.". How can I check this?.

                    - Because the clustering has failed, there are no thumbnails to review, except for one run with 64 k/mm2. The thumbanails shows almost no signal. Overclustering is not the issue.

                    Any thoughts are wellcome, and thank you so much for your help.

                    Comment

                    • josdegraaf
                      Member
                      • Mar 2010
                      • 33

                      #11
                      did you consider the naoh you use? is it fresh and < 1mM final concentration in your pool?
                      it could be (but does not explain no phix clusters) that the denaturation is insufficient, try heat denaturation in that case (high GC amplicons).

                      Comment

                      • Markiyan
                        Senior Member
                        • Sep 2010
                        • 126

                        #12
                        rkiyan : PS: there are some GC stretches in those reads - so please also check the thermal cycling performance of the instrument during the cluster generation.". How can I check this?.

                        Basically one need to check for 2 things:

                        1. The flowcell makes a good thermal contact with the Miseq's thermal block during the thermal cycling (no gaps between flowcell glass and thermal block!)

                        2. Get a thermocouple to the side of the thermal block (or a thermal testing flow cell) and log the temperature changes during the cluster generation.

                        (Similar to: http://jamimmunology.blogspot.co.uk/...le-logger.html)

                        Assuming that the instrument is on the service contract - the Illumina's FAS should do the above steps.

                        - Because the clustering has failed, there are no thumbnails to review, except for one run with 64 k/mm2. The thumbanails shows almost no signal. Overclustering is not the issue.

                        Any thoughts are welcome, and thank you so much for your help.[/QUOTE]

                        And if you set up a pure PhiX run, does it work as expected?

                        If yes - something in your library is inhibiting the cluster formation...

                        PS: Any attempts to use a heat treatment step for the library denaturation before loading?

                        Comment

                        • GW_OK
                          Senior Member
                          • Sep 2009
                          • 411

                          #13
                          If there's no mention of custom primers in the sample sheet then that's not something to worry about.

                          My next thought would be to check the concentration of your NaOH. A heat denaturation attempt might be worthwhile, as several others have previously mentioned.

                          You could also just assume that there's some sort of inhibitor in there and do another cleanup of your library to try and remove whatever it is.

                          Comment

                          • bsvega
                            Member
                            • Jul 2015
                            • 10

                            #14
                            The libraires have two rounds of ampure XP cleaning. We also try the Generead size selection columns.

                            We check NaOH (same NaOH vial is working with other 4nM libraries). Maybe try higher concentrations of NaOH (need to neutralize before loading) and also heat denaturation.

                            Meanwhile I am going to try the Kappa kit with some samples. I am trying to solve the issue because there is no more DNA for some samples to repeat library (need to rescue them).

                            Thanks for all your thoughts.

                            Comment

                            • bsvega
                              Member
                              • Jul 2015
                              • 10

                              #15
                              Finally the problem was solved. The libraries were all OK, It was a problem with the Miseq not taking the sequencing primer!!!.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                10-30-2026, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              45 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              27 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              76 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              49 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...