  • Discrepancy between Qbit quant & read count - possible contamination?

    We are performing MinION sequencing of DNA extracted from fresh water samples.
    We got some preliminary results that suggested possible contamination, so in our most recent round of sequencing we also sequenced a reagent blank from the DNA extraction kit to see if we could rule out contamination from that step.

    We first used a Qbit to quant the DNA from each sample, and were encouraged by the result (DNA in the reagent blank was below the detectable threshold).

    We then prepped and sequenced our samples (and the blank) using the ONT Rapid Barcoding Kit 96, and found about 5500 reads containing the barcode assigned to the blank. This was a small fraction of the read counts we were seeing for our samples. We had other un-used barcodes that had reads, but never more than a dozen or so.

    Is it possible that we somehow introduced contamination during the prep?
    We're struggling with how to explain why the Qbit showed no DNA, but the sequencing resulted 5500 reads for the same sample (they mostly classified as E. coli and human when run through the WIMP pipeline).

    Has anyone encountered this before? Any advice would be appreciated.

  • #2
    Have you added DCS to your samples during library prep? The human component could be contamination but the E.coli reads might be alignments to the DCS (lambda DNA).

    Comment


    • #3
      Hello Anmar06, I have the same question!
      backrooms game

      Comment


      • #4
        Thanks for the suggestion.

        For the library prep itself, we only added the reagents in the ONT Rapid sequencing Kit (following the ONT protocol). We did not include the lambda DNA when running our samples.

        We had run the lambda DNA prep and sequencing when we first set up the MinION, but on an entirely different flow cell from any of our samples.

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Current Approaches to Protein Sequencing
          by seqadmin


          Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
          04-04-2024, 04:25 PM
        • seqadmin
          Strategies for Sequencing Challenging Samples
          by seqadmin


          Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
          03-22-2024, 06:39 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, 04-11-2024, 12:08 PM
        0 responses
        31 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 10:19 PM
        0 responses
        33 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 09:21 AM
        0 responses
        28 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-04-2024, 09:00 AM
        0 responses
        53 views
        0 likes
        Last Post seqadmin  
        Working...
        X