    Using the illumina mRNA Seq kits I created 8 libraries last week; they are completed through pcr enrichment and ready to run on the GAs. For the most part the traces looked good however on all 8 samples there is an odd peak at around 75 bps. When I performed a gel cut out at ~300 bps I don't believe I introduced any other material so I am not sure where the extra peak is coming from. It seems to me that it could be a primer dimer from pcr enrichment or some sort of contamination in the gel of my bioanalyzer kit, either way I don't think those would affect the reads on a GA but I'd like to hear some input from you guys. The scans are attached bellow
    Attached Files
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    If you used a TruSeq kit to construct these, then my guess is that the ~75 "bps" peak is actually the PCR primers used for enrichment PCR. These primers, from the "PPC" (PCR primer cocktail) tube give a peak in that region of the electropherogram of both DNA high sensitivity chips and RNA (pico) chips. See:

    Any non-primary sequence heritable modification of genetic material. ChIP-SEQ, DNA methylation (Bisulfite-SEQ), chromatin modifications (methylation, acetylation, etc), non coding RNA.



    We don't run the DNA1000 chips in our lab, so have never looked at it on that type of chip.

    --
    Phillip

    Comment

    Latest Articles

    Collapse

    • SEQadmin2
      Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
      by SEQadmin2



      Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
      ...
      Yesterday, 11:10 AM
    • SEQadmin2
      Cancer Drug Resistance: The Lingering Barrier to Rising Survival
      by SEQadmin2



      Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

      There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
      07-08-2026, 05:17 AM
    • GATTACAT
      Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
      by GATTACAT
      Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
      07-01-2026, 11:43 AM

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by SEQadmin2, Yesterday, 10:04 AM
    0 responses
    11 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 07-08-2026, 10:08 AM
    0 responses
    9 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 07-07-2026, 11:05 AM
    0 responses
    16 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 07-02-2026, 11:08 AM
    0 responses
    31 views
    0 reactions
    Last Post SEQadmin2  
    Working...