Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • blair.chr
    Junior Member
    • Mar 2013
    • 4

    #1

    ddRAD-Seq

    Hi all,

    I am following the ddRAD-Seq protocol of Peterson et al. 2012. I start with ~40-100 ng/ul DNA as measured on a nanodrop. I take 5 ul of this and double digest with two different enzymes. I then clean this entire product with Ampure XP beads.

    The problem is that I am losing A LOT of DNA at every step. If I run the uncleaned digests on a bioanalyzer I get about 3-9 ng/ul. After Ampure cleanup my concentrations decrease to less than 1 ng/ul. As the standard protocols for restriction digests and bead cleanup are fairly straightforward, I am not sure where the problem lies. Any thoughts would be appreciated.

    Chris
  • ZWB
    Member
    • Apr 2011
    • 27

    #2
    Hi Chris,

    We've had similar problems with determining concentrations. The problem may actually be the nanodrop rather than your protocol. We switched to a Qubit and get much more reliable results. We found that any impurities in your sample really throw off the nanodrop, but using a fluorescence based dye as with the Qubit fixes that issue. Good luck.

    Comment

    • srdoyle
      Junior Member
      • Sep 2011
      • 1

      #3
      Hi Chris,

      Just wanted to check in and see how your ddRAD experience is going... we might be having similar issues and am keep to see how you dealt with them. Happy to chat via email too if you have some time.

      Cheers,
      Steve

      Comment

      • horbachc
        Junior Member
        • May 2013
        • 2

        #4
        What platform are you running your ddRAD on? I am using MiSeq, and have had issues with low cluster density compared to other library prep protocols (Nextera XT) we have used. What concentration are you loading?

        Comment

        • maxbangs
          Member
          • Apr 2014
          • 10

          #5
          Hello blair.ch, I do not know if you have solved your problem or not, I know this response is very late. I have just started with the Peterson protocol and may have a little help. Please note I am a complete novice (started two weeks ago) and may be way off base.

          First off if you loaded 5ul of 40-100ng/ul DNA into a standard 50ul reaction then your concentration should be 1/10th of what you started with, so 4-10ng/ul. This would explain why you got 3-9ng/ul concentrations for your digest.

          Second with the AMPure cleanup only DNA fragments of >100bps will attach to the beads, thus if most of you digest is <100bps then this may explain why the DNA disappeared after the clean up. I had this problem with a couple of samples. These samples looked great on the Qubit (>1000ng/ul of DNA) but when I looked at the quality of the DNA on a gel the sample was highly degraded and lacked any intact genome DNA. Thus when I looked at the digest on the gel all of the DNA fragments were <100bps.

          Hope this helps, also I am curious if you figured out your problem and had any advice for us just starting.

          Comment

          • rosatoc
            Member
            • Apr 2008
            • 22

            #6
            Hi. I have just tried my first double digest (PstI/MspI) for library construction, but using the GBS method rather than the RAD-Seq. On wheat. The Bioanalyzer trace has a huge adapter dimer and a bit of other 'stuff' which is too close to baseline to interpret. With those disappointing results we ran a gel following a digest to see whether the sample(s) are restrictible, and though we see a smear there is evidence of a lot of high MW DNA which may not have been cut. Are there any suggestions on what to try next to get this method to work? We're using columns for clean-up rather than beads, and we may switch though I don't think that would make a difference here.

            Many thanks for any tips, tricks, etc. you are able to share

            Comment

            • nucacidhunter
              Jafar Jabbari
              • Jan 2013
              • 1250

              #7
              I suspect you have followed Poland et al protocol. I may be able to provided some suggestions. Could you post your Bioanalyser trace (is it library after PCR clean up?) and your digest gel photo with a bit of details (DNA input, duration of digestion and loaded amount). It is usual not to see much of smaller fragments. If column is used for clean up, a titration with different amount of adapters is essential, but one can save a lot of time just by doing a 1.2x bead clean up which will remove excess adapters and relatively short fragments which does not provided useful data.

              Comment

              • maxbangs
                Member
                • Apr 2014
                • 10

                #8
                I am also using Pst I x MspI on some fish with a genome size of ~2.42Gbps. I attached a gel picture of a successful digest, the ladder is 1kb, 750bp, 500bp, 300bp, 100bp, 50bp. I used 10units for each RE and 1000ng of DNA in a 50ul reaction ran over night (17hours). Though this may be helpful to the discussion since it sounds like you have a problem with the digest.
                Attached Files
                Last edited by maxbangs; 06-05-2014, 05:46 PM.

                Comment

                • SNPsaurus
                  Registered Vendor
                  • May 2013
                  • 525

                  #9
                  One issue with wheat is that PstI will very rarely cut since most of the genome is methylated, but MspI will cut the methylated DNA. I don't know how you calculated the adapter amounts to add, but the PstI adapter will only find overhangs at a very low rate compared to what is expected from the genome size.
                  Providing nextRAD genotyping and PacBio sequencing services. http://snpsaurus.com

                  Comment

                  • rosatoc
                    Member
                    • Apr 2008
                    • 22

                    #10
                    Thanks for all responses. As we are trying this using the Poland et al protocol, developed for wheat (and barley), we used the adapter concentrations according to the protocol without a titration. That may be something we will need to consider. We have a few ideas about how to do the next test library construction and we feel success is right around the corner.... Hope springs eternal.

                    Thanks, again, for all replies

                    Comment

                    • rosatoc
                      Member
                      • Apr 2008
                      • 22

                      #11
                      Hi - we are still trying to get the double digest to work in our lab. We know that the digest is working (gel imaging) but the ligation results in adapter dimer peak only post PCR. There is a heat inactivation step following the digest with PstI-HF/MspI, however neither of these enzymes are heat inactivated.... Is there any concern that the active enzymes will prevent ligation to occur?

                      Comment

                      Latest Articles

                      Collapse

                      • SEQadmin2
                        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                        by SEQadmin2



                        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                        ...
                        10-14-2026, 09:55 AM
                      • SEQadmin2
                        New Genomics Technologies Take Aim at Long-Standing Limits
                        by SEQadmin2


                        Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                        We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                        ...
                        09-28-2026, 10:25 AM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, 10-22-2026, 11:13 AM
                      0 responses
                      37 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-22-2026, 10:48 AM
                      0 responses
                      28 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-15-2026, 10:52 AM
                      0 responses
                      41 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-14-2026, 05:39 AM
                      0 responses
                      45 views
                      0 reactions
                      Last Post SEQadmin2  
                      Working...