    Ampure bead clean up

    Hello,
    How much drying time do you normally allow when using ampure beads XP to clean up your library preps?

    I've been using the standard 15 minutes as mentioned in the TruSeq DNA library prep method (or less if the beads appear to be drying quicker). However, I have recently heard of other labs using just 2 minutes drying time as they believe the beads should be a little wet and sticky.

    What does everyone think?

    Thanks for your help,
    Anna.
  • chevrm
    Member
    • Apr 2012
    • 14

    #2
    15 minutes is a bit overkill. You don't want your beads to "crack" when drying, but you don't want them to have residual ethanol either (effects elution). General rule I use is 2-5 minutes, but just give your plate/tube an eyeball and make sure they aren't wet.

    Comment

    • anna_m
      Member
      • Jun 2013
      • 17

      #3
      Hi,
      Thanks for your reply. 2 minutes just doesn't seem enough. Is there not any ethanol carryover this way? I keep an eye on the beads and usually 10-15 minutes is fine so that the beads appear matt and not shiny.

      Comment

      • sara_ahmed
        Member
        • Nov 2011
        • 13

        #4
        Yes, 2 mins at room temperature is generally not enough. You can put them in the speedvac for a couple of mins to speed up the drying.

        Comment

        • chevrm
          Member
          • Apr 2012
          • 14

          #5
          I failed to mention that it depends greatly on what type of plate/tube you're using. In a V-bottom plate, it will take a few minutes longer than a U-bottom because of the more-gradual taper. I agree with the speedvac suggestion if you are using a tubebased approach.

          Comment

          • snorberg
            Junior Member
            • Jan 2013
            • 5

            #6
            I use 15m as well, but just looked over the manufacture's protocol and found this note...

            NOTE: A dry time of >5 min at Room Temperature is optional to ensure all traces of Ethanolare removed but take care not to over dry the bead ring (bead ring appears cracked) as this will significantly decrease elution efficiency.After a visual inspection, I resuspend the beads after 10 minutes.

            ...so maybe 5 minutes would be safer?

            Comment

            • JamieHeather
              @jamimmunology
              • Nov 2012
              • 96

              #7
              Sorry snorberg, but that's a slight (but crucial!) misquote:

              A dry time of ≤5 min at Room Temperature is optional to ensure all traces of Ethanol are removed but take care not to over dry the bead ring (bead ring appears cracked) as this will significantly decrease elution efficiency.

              Comment

              • EvelyneLuyten
                Junior Member
                • Jun 2013
                • 5

                #8
                Manually we do it for 15minutes, because if there is still ethanol left, the quality of the library is not good.
                With our Sciclone, it's 2-5 minutes on 37°C.
                I don't think it's bad when it's cracked, but you must see that the beads won't 'fly away' eg. with use under a flow.

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                  by SEQadmin2



                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                  ...
                  10-14-2026, 09:55 AM
                • SEQadmin2
                  New Genomics Technologies Take Aim at Long-Standing Limits
                  by SEQadmin2


                  Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                  We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                  ...
                  09-28-2026, 10:25 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, 10-22-2026, 11:13 AM
                0 responses
                35 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 10:48 AM
                0 responses
                25 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-15-2026, 10:52 AM
                0 responses
                41 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-14-2026, 05:39 AM
                0 responses
                45 views
                0 reactions
                Last Post SEQadmin2  
                Working...