  • mark_6060
    Junior Member
    • Feb 2023
    • 7

    #1

    Tapestation profile weird

    Hi all,

    We are outsourcing RNA-seq along with RNA isolation from a gram positive bacterium. The service provider has sent the QC report (Tapestation) for the isolated RNA.
    The first sample is said to have failed and has the following profile.
    Click image for larger version

Name:	4.png
Views:	701
Size:	52.8 KB
ID:	323847

    For other sample, the RNA isolation is being claimed to have succeeded with the following profiles.

    Click image for larger version

Name:	1.png
Views:	869
Size:	48.0 KB
ID:	323848
    Click image for larger version

Name:	2.png
Views:	717
Size:	59.0 KB
ID:	323849
    Click image for larger version

Name:	3.png
Views:	874
Size:	57.5 KB
ID:	323850


    Do you have any suggestions regarding the following points?
    (1) Why there is weird peak between 25 and 200 and also a corresponding darkest band in all the samples?
    (2) Shall the service provider be asked to proceed with the library preperation for the bottom 3 samples?

    Many thanks
    Attached Files
  • Ben3
    Member
    • Sep 2022
    • 79

    #2
    mark_6060 it's been a while since I've done any bacterial work or run that Tapestation, but to me, it looks like the RNA is a bit degraded and that's why there are large peaks at the end.

    Do you have any older Tapestation runs to compare this to? Also, I remembered that for RNA Tapestation runs, the RNA has to be put into a thermocycler for denaturation prior to running on the screen tape. If the temperature is wrong it can degrade the RNA or cause it to run strangely on the screen tape.

    Depending on your application, slightly degraded RNA might not be a huge deal. But first I would look at other runs and see how they compare first.

    Comment

    • mark_6060
      Junior Member
      • Feb 2023
      • 7

      #3
      Ben3, thanks for your reply. These samples were analyzed for the first time; so I don;t have their earlier profiles. However, I do have profiles (pasted below) for some other samples that have passed the QC on the current batch.

      Click image for larger version

Name:	5.png
Views:	758
Size:	55.1 KB
ID:	323867

      Comment

      • Ben3
        Member
        • Sep 2022
        • 79

        #4
        mark_6060 those profiles look much better. That's more of what I expect out of quality RNA. Seeing how they compare, I would say the RNA from the first charts is slightly degraded, but not that bad. I think the question would be how comfortable you would feel if there was some slight degradation.

        Depending on the library prep kit and the aim of the study, it might not matter too much if there's a small bit of degradation.

        Comment

        • AngelaT
          Junior Member
          • Jul 2024
          • 7

          #5
          I realize this post is older, but I just had what seems like the same problem, so I figured I'd comment anyway.

          I'm not familiar with RNA, so I can't comment on the sample graphs themselves.

          I have had samples fail QC, even though they had very large peaks. What was happening is that the sample concentration was so high that it was stronger than the ladder, so the software was not able to pick up the ladder peaks. Try diluting the samples a bit and trying them again.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          22 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          19 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          36 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          42 views
          0 reactions
          Last Post SEQadmin2  
          Working...