    TapeStation 2200: multiple problems

    Dear users,

    this is my first post at SEQanswers. I work at a NGS core facility as scientific head, which involves a lot of troubleshooting.

    I hope to get some pointers for troubleshooting of our TapeStation. Agilent tried very hard to provide us with answers, but we do not think that any of them helped us locate the real problems and fix them.
    We measure mainly DNA samples such as libraries for NGS (HaloPlex, Shotgun) and gDNA from human tissue or blood.

    1. Already recognized tapes are rejected after measurement and the data is lost. This occurs up to four times in a row. (Barcodes are intact, the scanner is cleaned, the lid closed to avoid reflections).

    2. Single samples are not detected, but are measured normally upon reloading. There are no visible defects like air bubbles in the tapes. (We also excluded buffer composition, sample concentration).

    2. The sample cannot be calculated because the marker is not recognized. The marker is clearly visible in the plate due to its color. It even seems to run after the sample, which should not be possible.

    In all cases different kinds of tapes are affected (genomic, high sensitivity, 1000bp).

    We suspected a software problem, but according to Agilent our software is up to date. Could it be that something else is affecting the software or do you think our problems are a hardware problem?

    I would very much appreciate your thoughts on the matter.

    Best regards
    DarkStar

Latest Articles

Collapse

  • GATTACAT
    Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
    by GATTACAT
    Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
    07-01-2026, 11:43 AM
  • SEQadmin2
    Nine Things a Sample Prep Scientist Thinks About Before Sequencing
    by SEQadmin2


    I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

    Here are nine questions we think about, in roughly the order they matter, before...
    06-18-2026, 07:11 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, Yesterday, 11:08 AM
0 responses
6 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 06-30-2026, 05:37 AM
0 responses
11 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 06-26-2026, 11:10 AM
0 responses
19 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 06-17-2026, 06:09 AM
0 responses
53 views
0 reactions
Last Post SEQadmin2  
Working...