  • PrimerBuffalo
    Junior Member
    • Mar 2017
    • 6

    Variance in cDNA amplification

    Hello,

    I recently generated cDNA from 100 T-cells (from the same source) using the Smart-Seq2 protocol from Sandberg et al. After the pre-amplification step (all wells were pre-amplified for 18 cycles), I ran 1uL of my cDNA on a Bioanalyzer and noticed that the amount of cDNA that I recovered was relatively variable (range from ~450pg/uL to ~1,400pg/uL).

    Does anyone know how much variance in cDNA concentration is expected well-to-well? Is this variance biological or an artifact of the protocol?

    I would like to get an idea I am worried that the variance that I observe is due to technical error on my part, perhaps in pipetting.

    Thanks in advance for the help!
  • Simone78
    Senior Member
    • Oct 2010
    • 208

    #2
    Originally posted by PrimerBuffalo View Post
    Hello,

    I recently generated cDNA from 100 T-cells (from the same source) using the Smart-Seq2 protocol from Sandberg et al. After the pre-amplification step (all wells were pre-amplified for 18 cycles), I ran 1uL of my cDNA on a Bioanalyzer and noticed that the amount of cDNA that I recovered was relatively variable (range from ~450pg/uL to ~1,400pg/uL).

    Does anyone know how much variance in cDNA concentration is expected well-to-well? Is this variance biological or an artifact of the protocol?

    I would like to get an idea I am worried that the variance that I observe is due to technical error on my part, perhaps in pipetting.

    Thanks in advance for the help!
    Hi,
    the variation is prefectly within the range we normally see when running SS2. I believe the difference is partly biological (cells have different sizes and hence different amount of RNA, are in different parts of the cell cycle, etc) and partly due to an artifact (PCR bias). If you BA profile looks fine (nice peak around 1.5-2 kb, no or little primer dimers, low "background") I wouldn´t worry!
    /Simone

    Comment

    • PrimerBuffalo
      Junior Member
      • Mar 2017
      • 6

      #3
      Thank you so much! That is very reassuring!

      Comment

      • PrimerBuffalo
        Junior Member
        • Mar 2017
        • 6

        #4
        Originally posted by Simone78 View Post
        Hi,
        the variation is prefectly within the range we normally see when running SS2. I believe the difference is partly biological (cells have different sizes and hence different amount of RNA, are in different parts of the cell cycle, etc) and partly due to an artifact (PCR bias). If you BA profile looks fine (nice peak around 1.5-2 kb, no or little primer dimers, low "background") I wouldn´t worry!
        /Simone
        Hi Simone78,

        I am currently generating indexed libraries from my samples and have run into a problem:

        I used 1ng of cDNA and ran the indexing PCR for 7 cycles. I did this because I thought using more starting cDNA would result in more product, but that seems not the case.... By bioanalyzer I ended up with too low a concentration of indexed cDNA (200-1000 pMolar) and my avg peak size was 800-1000 bp.

        So I feel that I must lower the amount of starting cDNA, and increase the number of cycles.
        In your original protocol you use 0.1 ng cDNA for 12 cycles. My question is why do you use so little cDNA when you are probably generating much more during the preamplification step. Why not use 0.5 ng etc?


        Thanks for your help!

        Comment

        • Simone78
          Senior Member
          • Oct 2010
          • 208

          #5
          Originally posted by PrimerBuffalo View Post
          Hi Simone78,

          I am currently generating indexed libraries from my samples and have run into a problem:

          I used 1ng of cDNA and ran the indexing PCR for 7 cycles. I did this because I thought using more starting cDNA would result in more product, but that seems not the case.... By bioanalyzer I ended up with too low a concentration of indexed cDNA (200-1000 pMolar) and my avg peak size was 800-1000 bp.

          So I feel that I must lower the amount of starting cDNA, and increase the number of cycles.
          In your original protocol you use 0.1 ng cDNA for 12 cycles. My question is why do you use so little cDNA when you are probably generating much more during the preamplification step. Why not use 0.5 ng etc?


          Thanks for your help!
          Hi,
          yes, probably 1 ng input is too much for the Tn5 you add, especially when using the Nextera XT kit. This leads to under-tagmentation and libraries with long(er) average size. Moreover, I would use 10 PCR cycles when starting from 1 ng.
          You are also correct when you say that we should generate more cDNA than what we actually use. The reason why I start from 100 pg is just convenience. As shown on Fig 5D of our Genome Res paper, there is no difference in gene detection and SD when starting from 1 ng, 500 pg or 100 pg. By using less input I can therefore use less Tn5 (usually 0.1 ul of a 12.5 uM solution is sufficient) and make my batch last longer!
          I think that, at least when using Nextera XT, you should stay in the range of 100-500 pg. When I use the kit I generally do the reaction in 2 ul and use only 50-150 pg DNA and 0.5 ul Tn5. I hope this clarifies things, let me know otherwise.
          Best,
          Simone

          Comment

          • PrimerBuffalo
            Junior Member
            • Mar 2017
            • 6

            #6
            Originally posted by Simone78 View Post
            Hi,
            yes, probably 1 ng input is too much for the Tn5 you add, especially when using the Nextera XT kit. This leads to under-tagmentation and libraries with long(er) average size. Moreover, I would use 10 PCR cycles when starting from 1 ng.
            You are also correct when you say that we should generate more cDNA than what we actually use. The reason why I start from 100 pg is just convenience. As shown on Fig 5D of our Genome Res paper, there is no difference in gene detection and SD when starting from 1 ng, 500 pg or 100 pg. By using less input I can therefore use less Tn5 (usually 0.1 ul of a 12.5 uM solution is sufficient) and make my batch last longer!
            I think that, at least when using Nextera XT, you should stay in the range of 100-500 pg. When I use the kit I generally do the reaction in 2 ul and use only 50-150 pg DNA and 0.5 ul Tn5. I hope this clarifies things, let me know otherwise.
            Best,
            Simone
            Yes, that is incredibly helpful thank you!

            I will definitely reduce my input concentration and increase my number of cycles. I have a colleague who uses 12 cycles for an input of 0.5 ng and I noticed that you also use 12 cycles but for 0.1ng. Is there much difference in the number of cycles past 10?

            Comment

            • Simone78
              Senior Member
              • Oct 2010
              • 208

              #7
              Originally posted by PrimerBuffalo View Post
              Yes, that is incredibly helpful thank you!

              I will definitely reduce my input concentration and increase my number of cycles. I have a colleague who uses 12 cycles for an input of 0.5 ng and I noticed that you also use 12 cycles but for 0.1ng. Is there much difference in the number of cycles past 10?
              Hi,
              I don´t think it really matters if you do 10 or 12 cycles. The largest bias, where you "lost" your transcripts (due to many reasons that would take too long to explain here) is anyway introduced after the 18, 20 or more cycles you do in the first PCR (after RT). After the tagmentation the library is more homogeneous in terms of size, which should decrease the risk of preferential amplification of the short(er) fragments. Sensitivity and reproducibility is more a matter of input used in the tagmentation. If you use too little DNA you might end up with a sub-population of transcripts that is not representative of your real sample and which has low complexity.
              Best,
              Simone

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                by SEQadmin2


                I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                Here are nine questions we think about, in roughly the order they matter, before...
                06-18-2026, 07:11 AM
              • SEQadmin2
                From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                by SEQadmin2


                Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                ...
                06-02-2026, 10:05 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 06-17-2026, 06:09 AM
              0 responses
              40 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-09-2026, 11:58 AM
              0 responses
              102 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-05-2026, 10:09 AM
              0 responses
              123 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-04-2026, 08:59 AM
              0 responses
              114 views
              0 reactions
              Last Post SEQadmin2  
              Working...