    Two cDNA Peaks on Bioanalyzer

    I used 200 ng total RNA input and the KAPA HyperPrep Kit to generate cDNA libraries. Usually, our libraries look like Sample 1 in the attached pdf, but this time we saw two peaks and a trailing 'belly'. Does anyone have an idea what this could be? Of note, for this High-Sensitivity Bioanalyzer run, I left the cDNA samples at room temperature for 30 min, together with the Bioanalyzer reagents; I usually leave the samples on ice, hence I mention it; maybe this could be a reason? Otherwise, I am thinking of overamplification, i.e. too many cycles for the library amplification step. It would be great to get some feedback and advice on this.
    Attached Files
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    Hi patkrat,
    Yes, the apparently longer products are likely "bubble products". It is hypothesized that bubble products form during PCR when the concentration of denatured products is high enough to allow strands from different library molecules to anneal at their adapters, leaving the two insert strands unannealed because they are not related. The floppy insert strands apparently retard the progress of these bubble product through the sieving matrix of the chip.

    You can ignore the issue if you like. The bubble products will denature along with the normal double stranded products during denaturation and cluster normally. But if you need an average size for your library to calculate its concentration (eg, for qPCR), then you want to use the left peak to determine that.

    --
    Phillip

    Comment

    • patkrat
      Member
      • Apr 2017
      • 14

      #3
      Originally posted by pmiguel View Post
      Hi patkrat,
      Yes, the apparently longer products are likely "bubble products". It is hypothesized that bubble products form during PCR when the concentration of denatured products is high enough to allow strands from different library molecules to anneal at their adapters, leaving the two insert strands unannealed because they are not related. The floppy insert strands apparently retard the progress of these bubble product through the sieving matrix of the chip.

      You can ignore the issue if you like. The bubble products will denature along with the normal double stranded products during denaturation and cluster normally. But if you need an average size for your library to calculate its concentration (eg, for qPCR), then you want to use the left peak to determine that.

      --
      Phillip
      Thanks for your response, Phillip. In fact, I ran another HighSens chip with 1:10 dilution of the cDNA library and found that this completely resolved the issue. While you may be right about the bubble product, I think that the main issue in this case actually was simple overloading of the chip (which has a 10 ng max capacity); I sometimes do a quick photospec to get the library concentration before running it on the Bioanalyzer, but this time the concentration was way off. I checked with the Qubit and found that I actually had up to 30 ng/ul libraries.

      Conclusion: Overloading of the chip causes a split peak, but 30 ng/ul also indicates too many library amp cycles, which I will reduce from 13 to 12 (or maybe even 11) next time.

      Thanks again for your answer.

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        Thanks for that follow-up. Interesting.

        --
        Phillip

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM
        • SEQadmin2
          New Genomics Technologies Take Aim at Long-Standing Limits
          by SEQadmin2


          Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

          We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
          ...
          09-28-2026, 10:25 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        47 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        35 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-15-2026, 10:52 AM
        0 responses
        46 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-14-2026, 05:39 AM
        0 responses
        46 views
        0 reactions
        Last Post SEQadmin2  
        Working...