    ddRAD Post-PCR Bioanalyzer Has Multiple Peaks

    Hello,

    I have a problem with a post-PCR ddRAD library that was prepared using Peterson (2012). It was prepared along with another index for sequencing on one Illumina lane, and the pre-PCR size selection for both indices was supposed to be 454-509 bp.

    However, the Bioanalyzer results show two smaller peaks for one of the indices that shouldn't be there (see attached images), and the smaller peaks should be outside the size-selection range. Thus, it must have post size selection.

    Does anyone know what might have caused the two smaller peaks to occur, and should I worry about it during sequencing?

    Thank you,

    -Bradley
    Attached Files
  • BioKiwi
    Member
    • Sep 2014
    • 19

    #2
    I wonder how many PCR cycles you have done and what was the total yield.

    Comment

    • evobio721
      Junior Member
      • Dec 2017
      • 5

      #3
      Thank you for your prompt reply. I did 12 PCR cycles. The post-PCR yield was 45.6 ng/uL, which come to think of it does seem high. Pre-PCR DNA was ~1 ng/uL.

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        If you have run size selected tags on BA or still have remaining to run you can check for the presence of those peaks. I think the size selection has been sub optimal.

        Other weak possibility is over amplification. I wonder what was the PCR input and final library elution volumes.

        If it turns out to be over-amplification then a new library needs to be prepped. If it is result of size selection then you just will loose some reads and library can be sequenced.

        Comment

        • evobio721
          Junior Member
          • Dec 2017
          • 5

          #5
          Originally posted by nucacidhunter View Post
          If you have run size selected tags on BA or still have remaining to run you can check for the presence of those peaks. I think the size selection has been sub optimal.

          Other weak possibility is over amplification. I wonder what was the PCR input and final library elution volumes.

          If it turns out to be over-amplification then a new library needs to be prepped. If it is result of size selection then you just will loose some reads and library can be sequenced.
          Thanks for your input. The size selection was done on a Pippin Prep, and the elution volume was what the Pippin Prep manual recommends (can't remember exact amount at the moment, but I think 40uL?). PCR input was 5 uL repeated four times, followed by post-PCR AMPure XP pooling/cleanup of PCR product.

          That's a good idea about running the size selected DNA on the BA. I will try that to see if the issue is pre or post PCR. I imagine that's about the only thing I can do at this point outside of re-doing library prep.

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            So you have used in total 20 ng of size selected tags for PCRs and got 45.6 ng/ul after pooled PCR clean up. I am interested to know what was the elution volume from beads to look at possibility of over-amplification.

            Comment

            • evobio721
              Junior Member
              • Dec 2017
              • 5

              #7
              Originally posted by nucacidhunter View Post
              So you have used in total 20 ng of size selected tags for PCRs and got 45.6 ng/ul after pooled PCR clean up. I am interested to know what was the elution volume from beads to look at possibility of over-amplification.
              After the post-PCR AMPure cleanup, I eluted in 30 uL. So there was 45.6 ng/uL in 30 uL.

              I should note that when I did the PCR, both the index that worked well (first image) and the index that contained the multiple peaks were done at the same time.

              Comment

              • nucacidhunter
                Jafar Jabbari
                • Jan 2013
                • 1250

                #8
                It does not seem to be result of over-amplification. Most likely cassette lane has been faulty or other weak cause could be amplicon contamination from the lab if you have recently processed amplicons in those sizes recently in the same location.

                Comment

                • evobio721
                  Junior Member
                  • Dec 2017
                  • 5

                  #9
                  Originally posted by nucacidhunter View Post
                  It does not seem to be result of over-amplification. Most likely cassette lane has been faulty or other weak cause could be amplicon contamination from the lab if you have recently processed amplicons in those sizes recently in the same location.
                  That sounds reasonable to me. Thank you very much for your time and help. I greatly appreciate it.

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                    by SEQadmin2



                    Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                    ...
                    07-09-2026, 11:10 AM
                  • SEQadmin2
                    Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                    by SEQadmin2



                    Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                    There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                    07-08-2026, 05:17 AM
                  • GATTACAT
                    Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                    by GATTACAT
                    Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                    07-01-2026, 11:43 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, 07-13-2026, 10:26 AM
                  0 responses
                  22 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 07-09-2026, 10:04 AM
                  0 responses
                  32 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 07-08-2026, 10:08 AM
                  0 responses
                  20 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 07-07-2026, 11:05 AM
                  0 responses
                  34 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...