    Hello,

    We've been having some quantitation issues recently, I'm wondering whether the Qubit fluorometers are good for processing lots of samples (96 at a time?)
  • r.rosati
    Member
    • Aug 2015
    • 95

    #2
    Hi,
    they should be... Once you've passed the 2-minute minimum incubation time, then the tubes are supposed to be stable for 3 hours (Qubit dsDNA HS Assay guide, page 2), although personally it never happened that I had them them sit for more than half an hour prior to reading them, so I can't speak from personal experience.
    What kind of issues are you experiencing? Do you notice readings drop, or raise during time? Or do the readings just fluctuate? How do the calibrators perform at the end of the readings?
    Or is your issue that at the end of the experiment, the proportion of libraries is uneven? In this case, besides quantity, are they even in quality?

    Comment

    • microgirl123
      Senior Member
      • Jun 2012
      • 199

      #3
      What are you using for quantification now? I find the Qubit to be time-consuming for 96 samples - instead I use the Qubit reagents, white strip tubes, and a plate reader that can handle tubes.

      Comment

      • kmcarr
        Senior Member
        • May 2008
        • 1181

        #4
        Originally posted by microgirl123 View Post
        What are you using for quantification now? I find the Qubit to be time-consuming for 96 samples - instead I use the Qubit reagents, white strip tubes, and a plate reader that can handle tubes.
        microgirl,

        What plate reader are you using?

        Comment

        • microgirl123
          Senior Member
          • Jun 2012
          • 199

          #5
          Well. . . it's actually an old Stratagene qPCR instrument that will allow you to do a single endpoint read so it's only used as a "plate" reader now.

          I'm not sure how well Qubit reactions would work in a plate since there's not a good way to really vortex the samples, but I haven't experimented. I half the volume of the reaction in the strip tubes both to save reagent costs and to allow more room for mixing.

          Comment

          • ychang
            Member
            • Oct 2010
            • 17

            #6
            Hi,microgirl,

            your idea of using a qPCR instrument as a Qubit reader was great! could you give more detail of the protocol? Thanks!

            Comment

            • Carcharodon
              Member
              • Jul 2015
              • 40

              #7
              Originally posted by microgirl123 View Post
              Well. . . it's actually an old Stratagene qPCR instrument that will allow you to do a single endpoint read so it's only used as a "plate" reader now.

              I'm not sure how well Qubit reactions would work in a plate since there's not a good way to really vortex the samples, but I haven't experimented. I half the volume of the reaction in the strip tubes both to save reagent costs and to allow more room for mixing.
              I use the Biotium AccuClear system, and for that you simply mix the DNA with the dye/buffer mixture by pipetting and allow it to sit for ~ 5 mins. I would bet good money that the same approach would work with the Qubit reagents.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                by SEQadmin2


                Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                The systematic characterization of the human proteome has
                ...
                07-20-2026, 11:48 AM
              • SEQadmin2
                Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                by SEQadmin2



                Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                ...
                07-09-2026, 11:10 AM
              • SEQadmin2
                Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                by SEQadmin2



                Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                07-08-2026, 05:17 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 07-24-2026, 12:17 PM
              0 responses
              30 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-23-2026, 11:41 AM
              0 responses
              23 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-20-2026, 11:10 AM
              0 responses
              213 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-13-2026, 10:26 AM
              0 responses
              79 views
              0 reactions
              Last Post SEQadmin2  
              Working...