    #1

    Is Qubit good for multiple samples?

    Hello,

    We've been having some quantitation issues recently, I'm wondering whether the Qubit fluorometers are good for processing lots of samples (96 at a time?)
  • r.rosati
    Member
    • Aug 2015
    • 95

    #2
    Hi,
    they should be... Once you've passed the 2-minute minimum incubation time, then the tubes are supposed to be stable for 3 hours (Qubit dsDNA HS Assay guide, page 2), although personally it never happened that I had them them sit for more than half an hour prior to reading them, so I can't speak from personal experience.
    What kind of issues are you experiencing? Do you notice readings drop, or raise during time? Or do the readings just fluctuate? How do the calibrators perform at the end of the readings?
    Or is your issue that at the end of the experiment, the proportion of libraries is uneven? In this case, besides quantity, are they even in quality?

    Comment

    • microgirl123
      Senior Member
      • Jun 2012
      • 199

      #3
      What are you using for quantification now? I find the Qubit to be time-consuming for 96 samples - instead I use the Qubit reagents, white strip tubes, and a plate reader that can handle tubes.

      Comment

      • kmcarr
        Senior Member
        • May 2008
        • 1181

        #4
        Originally posted by microgirl123 View Post
        What are you using for quantification now? I find the Qubit to be time-consuming for 96 samples - instead I use the Qubit reagents, white strip tubes, and a plate reader that can handle tubes.
        microgirl,

        What plate reader are you using?

        Comment

        • microgirl123
          Senior Member
          • Jun 2012
          • 199

          #5
          Well. . . it's actually an old Stratagene qPCR instrument that will allow you to do a single endpoint read so it's only used as a "plate" reader now.

          I'm not sure how well Qubit reactions would work in a plate since there's not a good way to really vortex the samples, but I haven't experimented. I half the volume of the reaction in the strip tubes both to save reagent costs and to allow more room for mixing.

          Comment

          • ychang
            Member
            • Oct 2010
            • 17

            #6
            Hi,microgirl,

            your idea of using a qPCR instrument as a Qubit reader was great! could you give more detail of the protocol? Thanks!

            Comment

            • Carcharodon
              Member
              • Jul 2015
              • 40

              #7
              Originally posted by microgirl123 View Post
              Well. . . it's actually an old Stratagene qPCR instrument that will allow you to do a single endpoint read so it's only used as a "plate" reader now.

              I'm not sure how well Qubit reactions would work in a plate since there's not a good way to really vortex the samples, but I haven't experimented. I half the volume of the reaction in the strip tubes both to save reagent costs and to allow more room for mixing.
              I use the Biotium AccuClear system, and for that you simply mix the DNA with the dye/buffer mixture by pipetting and allow it to sit for ~ 5 mins. I would bet good money that the same approach would work with the Qubit reagents.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                by SEQadmin2




                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                10-30-2026, 06:46 AM
              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                10-14-2026, 09:55 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 10-28-2026, 09:57 AM
              0 responses
              38 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-28-2026, 09:51 AM
              0 responses
              23 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 11:13 AM
              0 responses
              72 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 10:48 AM
              0 responses
              48 views
              0 reactions
              Last Post SEQadmin2  
              Working...