    Yeast ChIP-seq; Sonication/Bioanalyzer question

    Hi,
    I've just started doing ChIP-seq, and am having difficulty getting my INPUT and IP DNA fragment sizes in the proper size range for Illumina sequencing. My ChIP protocol seems to work fine, as I can show enrichment of a 360bp region surrounding a known binding site in my IP sample compared to the INPUT sample by pcr.

    However, our Bioanalyzer reports that the size fragmentation is too small for the library prep. This seems to indicate that I'm over-sonicating the samples. I've been following a protocol published earlier this year by Mike Snyder's lab http://dx.doi.org/10.1016/S0076-6879(10)70004-5. In this protocol the sonication procedure uses a Branson Digital Sonifier @ 5x30sec @ power 50.

    In our lab, we have a Branson Analog Sonifier and I've tried sonicating 6x10sec @ power 20 and 3x10sec @ power 20. In both cases, the bioanalyzer indicates that my DNA size fragmentation is mostly <100bp. However, when I run the INPUT DNA on a 2% Agarose gel, I see a smear with the greatest DNA concentration ~1,000bp and no DNA below 100bp.
    Last edited by AaronS; 01-14-2011, 11:13 AM.
  • jorgemon
    Junior Member
    • Feb 2009
    • 4

    #2
    Hi AaronS.
    I think that the not real result is the agarose gel. The bioanalyzer is more specific than gel, where you have less resolution to see small DNA quantity and migration is not the same (the upper marker of a DNA HS is 10380bp, but you see all the largest fragmented sizes grouped around 1500bp). If you try with DNA 1000 or 7500, you will have the same result (you will see all large sizes grouped around 1500bp too).
    I think that the smear of fragmented sizes that you see in bioanalyzer, you can't see in agarose...you only can see sizes with more concentration (usually largest sizes but they are in correct size compared with marker, not as occurs in bioanalyzer).
    The sample that you see in bioanalyzer at 100bp, may be DNA degradation of Chip and Input. If you purify the samples by column (Qiaquick for example), and run again the samples at bioanalyzer, you will not see anything less than 150bp.
    You can try to fragment more Inputs with different times. Then you can run it in a bioanalyzer and know if you are fragment too much your samples.
    It can be a protein contamination from previous step too (it could explain this results).
    In order to can see better the DNA purified by column, you can put 2-3ul evaporated (quantitative range of DNA HS is up 0.5ng/ul).
    I hope you can solve this problem with my comments.
    Regards.

    Comment

    Latest Articles

    Collapse

    • SEQadmin2
      Nine Things a Sample Prep Scientist Thinks About Before Sequencing
      by SEQadmin2


      I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

      Here are nine questions we think about, in roughly the order they matter, before...
      06-18-2026, 07:11 AM
    • SEQadmin2
      From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
      by SEQadmin2


      Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


      The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
      ...
      06-02-2026, 10:05 AM

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by SEQadmin2, 06-26-2026, 11:10 AM
    0 responses
    12 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 06-17-2026, 06:09 AM
    0 responses
    48 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 06-09-2026, 11:58 AM
    0 responses
    107 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 06-05-2026, 10:09 AM
    0 responses
    125 views
    0 reactions
    Last Post SEQadmin2  
    Working...