    bioanalyzer and qubit disagreement

    We prepare clinical samples with a formal SOP and do not regularly see irregularities. Today we have a capture pool of samples that quanted at 24.0ng/uL on qubit. To determine fragment size we also run sample pool in duplicate on bioanalyzer DNA1000. Duplicate 1, there is noise around 40bp (<1ng/uL) and no peak in the expected range (300-400). Duplicate 2, there is noise around 40bp and a wide peak between 700 and 1500 (upper marker). We reran the sample pool (4 replicates) on the bioanalyzer with only the noise around 40bp.

    Qubit concentration = normal
    bioanalyzer fragment size = missing?

    Any ideas what is going on? Where is the missing peak?
    Last edited by smsestes; 02-05-2020, 02:47 PM. Reason: clarity
  • JakobHedegaard
    Member
    • Mar 2008
    • 62

    #2
    Perhaps you didn't load enough on the Bioanalyzer to get a signal from a smear using a D1000 chip? Try to run on HS DNA chip.

    Comment

    • Noveline
      Junior Member
      • Jan 2020
      • 4

      #3
      DNA 1000 Kit is for 5-50 ng/ul of DNA
      Hs DNA Kit: "NGS sheared DNA or libraries: For accurate determination of DNA concentration, the total DNA in the samples must be between 100pg/μL to 10 ng/μL."

      I think that in HS dna chip will be overloaded...

      Check if everything is ok with gel dye mix in DNA 1000 Kit.

      Comment

      • smsestes
        Junior Member
        • Feb 2020
        • 2

        #4
        Originally posted by Noveline View Post
        DNA 1000 Kit is for 5-50 ng/ul of DNA
        Hs DNA Kit: "NGS sheared DNA or libraries: For accurate determination of DNA concentration, the total DNA in the samples must be between 100pg/μL to 10 ng/μL."

        I think that in HS dna chip will be overloaded...

        Check if everything is ok with gel dye mix in DNA 1000 Kit.
        Hi Noveline,

        When we reran the sample pool, we included a previous sample pool which ran normally and as expected. Thus it is unlikely to be the BioA or its reagents.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
          by SEQadmin2


          Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

          The systematic characterization of the human proteome has
          ...
          07-20-2026, 11:48 AM
        • SEQadmin2
          Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
          by SEQadmin2



          Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
          ...
          07-09-2026, 11:10 AM
        • SEQadmin2
          Cancer Drug Resistance: The Lingering Barrier to Rising Survival
          by SEQadmin2



          Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

          There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
          07-08-2026, 05:17 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 07-20-2026, 11:10 AM
        0 responses
        18 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 07-13-2026, 10:26 AM
        0 responses
        32 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 07-09-2026, 10:04 AM
        0 responses
        43 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 07-08-2026, 10:08 AM
        0 responses
        29 views
        0 reactions
        Last Post SEQadmin2  
        Working...