  • smsestes
    Junior Member
    • Feb 2020
    • 2

    #1

    bioanalyzer and qubit disagreement

    We prepare clinical samples with a formal SOP and do not regularly see irregularities. Today we have a capture pool of samples that quanted at 24.0ng/uL on qubit. To determine fragment size we also run sample pool in duplicate on bioanalyzer DNA1000. Duplicate 1, there is noise around 40bp (<1ng/uL) and no peak in the expected range (300-400). Duplicate 2, there is noise around 40bp and a wide peak between 700 and 1500 (upper marker). We reran the sample pool (4 replicates) on the bioanalyzer with only the noise around 40bp.

    Qubit concentration = normal
    bioanalyzer fragment size = missing?

    Any ideas what is going on? Where is the missing peak?
    Last edited by smsestes; 02-05-2020, 02:47 PM. Reason: clarity
  • JakobHedegaard
    Member
    • Mar 2008
    • 62

    #2
    Perhaps you didn't load enough on the Bioanalyzer to get a signal from a smear using a D1000 chip? Try to run on HS DNA chip.

    Comment

    • Noveline
      Junior Member
      • Jan 2020
      • 4

      #3
      DNA 1000 Kit is for 5-50 ng/ul of DNA
      Hs DNA Kit: "NGS sheared DNA or libraries: For accurate determination of DNA concentration, the total DNA in the samples must be between 100pg/μL to 10 ng/μL."

      I think that in HS dna chip will be overloaded...

      Check if everything is ok with gel dye mix in DNA 1000 Kit.

      Comment

      • smsestes
        Junior Member
        • Feb 2020
        • 2

        #4
        Originally posted by Noveline View Post
        DNA 1000 Kit is for 5-50 ng/ul of DNA
        Hs DNA Kit: "NGS sheared DNA or libraries: For accurate determination of DNA concentration, the total DNA in the samples must be between 100pg/μL to 10 ng/μL."

        I think that in HS dna chip will be overloaded...

        Check if everything is ok with gel dye mix in DNA 1000 Kit.
        Hi Noveline,

        When we reran the sample pool, we included a previous sample pool which ran normally and as expected. Thus it is unlikely to be the BioA or its reagents.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
          by SEQadmin2



          CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

          Despite this, “CRISPR helped turn genome editing from a specialized technique into
          ...
          07-31-2026, 11:01 AM
        • SEQadmin2
          Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
          by SEQadmin2


          Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

          The systematic characterization of the human proteome has
          ...
          07-20-2026, 11:48 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, Today, 10:35 AM
        0 responses
        6 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 08-06-2026, 07:41 AM
        0 responses
        23 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 08-03-2026, 10:13 AM
        0 responses
        41 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 07-31-2026, 02:55 AM
        0 responses
        47 views
        0 reactions
        Last Post SEQadmin2  
        Working...