  • rougaroux
    Junior Member
    • Sep 2020
    • 3

    #1

    No amplification with Kapa kit

    Hello,
    I have been using the Kapa Hyper Prep kit to try and prep DNA libraries for target enrichment.
    I am using custom Y adapters and dual indexed primers which I know have been used successfully with the Kapa kit before.
    My issue is that I am getting virtually no amplification from the PCR. By increasing the number of cycles up to 16, my library concentrations increased by only 2-3X.


    The kit I'm working with is new.
    I checked my shearing with the bioanalyzer and all samples were within the targeted range (400-600 bp).

    I suspect the problem is with adapter ligation...after PCR, I see very high concentrations with a nanodrop (but not Qubit) that went down sharply after bead cleanup. This makes me think lots of adapter dimers and/or unused primers were present.

    Has anyone here experienced similar issues with the Kapa kit, or does anyone have any advice on how best to troubleshoot this problem?
  • luc
    Senior Member
    • Dec 2010
    • 469

    #2
    You could run aliquots of the reaction (before and after bead cleanups) on an agarose gel to see what is going on.
    If the ligation does not work porperly, perhaps your samples have some chemical contamination?

    Comment

    • rougaroux
      Junior Member
      • Sep 2020
      • 3

      #3
      Thanks for the suggestion, Luc.
      Do you know if I would be able to reliably identify unincorporated adapters on a gel?

      Comment

      • luc
        Senior Member
        • Dec 2010
        • 469

        #4
        Depends on the dye used and the amount of primers. Ethidiumbrmide is ten times less sensitive for ssDNA compared to dsDNA. The sampe seems to be true for the Bioanalyzer, but such an instrument is more sensitive than our eyes looking at an agarose gel.

        Comment

        • rougaroux
          Junior Member
          • Sep 2020
          • 3

          #5
          Okay, well I will give it a try with my post-ligation libraries before and after clean-up and see if there's anything to see.

          Does anyone have any other suggestions of common sources of library amplification failure?

          Comment

          • jdk787
            josh kinman
            • Apr 2014
            • 72

            #6
            If you're using Nanodrop on PCR products before cleanup, the "high concentration" you saw was likely just unincorporated nucleotides in the PCR mix, so that isn't useful.

            If you run your cleaned up post ligation product on the Bioanalyzer you should be able to see a shift to a larger size for your ligated products compared to your original fragmented DNA. Y adapters should make this shift more noticeable since libraries ligated with Y adapter migrate slower through the gel causing them to appear larger than their actual size.

            My best guess would be that there's something wrong with your custom adapters/primers or something went wrong with the prep. Maybe add a control sample that uses a non-custom adapter at ligation and p5/p7 primers for PCR.
            Last edited by jdk787; 10-03-2020, 05:43 AM.
            Josh Kinman

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
              by SEQadmin2



              CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

              Despite this, “CRISPR helped turn genome editing from a specialized technique into
              ...
              07-31-2026, 11:01 AM
            • SEQadmin2
              Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
              by SEQadmin2


              Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

              The systematic characterization of the human proteome has
              ...
              07-20-2026, 11:48 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 08-06-2026, 07:41 AM
            0 responses
            23 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 08-03-2026, 10:13 AM
            0 responses
            39 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-31-2026, 02:55 AM
            0 responses
            44 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-24-2026, 12:17 PM
            0 responses
            27 views
            0 reactions
            Last Post SEQadmin2  
            Working...