  • JPC
    Senior Member
    • May 2008
    • 116

    #1

    Expected Library size after adaptors

    If we get a pre adaptor, size selected library with an average size of 170 bp then after the adaptors (which are 41 and 53 bp) I would expect to see 170+41+53=264bp but what we actually see is more like 247bp. Does this seem normal? Perhaps the post adaptor size is not as expected because one of the adaptors is single stranded at it's 5' end?
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    What method are you using to size the library? Pre or post enrichment PCR?

    In any case I would not sweat the difference between 264 and 247 bp. Most likely the size is not exactly right. Keep in mind that if you are using some sort of electrophoresis based method that differences between the sequence composition of your sample and the size standard would likely be sufficient to explain a 7% difference. But other factors (salt concentration difference between the standard and sample) are also going to throw off your size estimation.

    --
    Phillip

    Comment

    • JPC
      Senior Member
      • May 2008
      • 116

      #3
      Thanks for your reply, we are using the Agilent Bioanalyser with the DNA 1000 kit and 'yes' we are aware of the variation that is inherent. It has since been pointed out to me that the bioanalyser only reads double stranded DNA and the Standard adaptor has over 30 bases of single strand so combining this with the variation seems to account for what we see.

      thanks again Phillip

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        Originally posted by JPC View Post
        It has since been pointed out to me that the bioanalyser only reads double stranded DNA
        Hi JPC,
        Not sure what was intended by the person that told you this.
        The bioanalyser DNA chips do "read" single stranded molecules. If by "read" you mean "visualize". You can check this yourself by running the ssRNA ladder for the RNA chips on a DNA chip. You will see peaks. Please see:

        Techniques and protocol discussions on sample preparation, library generation, methods and ideas


        for details. Counter-intuitively ssDNA segments would tend to actually slow the migration of molecules on the DNA agilient chips I have used.

        Actually, had you not already done the enrichment PCR step prior to your running your chip? If so, then the amplicon should be completely dsDNA.

        But I guess you library is fine in any case...

        --
        Phillip

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM
        • SEQadmin2
          New Genomics Technologies Take Aim at Long-Standing Limits
          by SEQadmin2


          Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

          We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
          ...
          09-28-2026, 10:25 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-15-2026, 10:52 AM
        0 responses
        15 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-14-2026, 05:39 AM
        0 responses
        21 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-13-2026, 09:56 AM
        0 responses
        20 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-09-2026, 11:14 AM
        0 responses
        27 views
        0 reactions
        Last Post SEQadmin2  
        Working...