Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • Rocketknight
    Member
    • Sep 2011
    • 86

    #1

    Cutting corners with Illumina Sample Prep

    Hey all, just a quick question regarding Illumina's TruSeq DNA sample prep protocol. After adapter ligation, there's a cleanup step with Ampure XP. If you're doing the gel-cut size-selection step though, that step comes directly afterwards. The manual seems to suggest that you need to do the Ampure cleanup even if you're planning to do the gel-cut afterwards, but I feel like electrophoresis should remove any contaminants well enough on its own.

    Has anyone tried skipping straight to electrophoresis without the Ampure cleanup? Is there any reason that shouldn't work?
  • ECO
    --Site Admin--
    • Oct 2007
    • 1360

    #2
    I just skip the electrophoresis altogether, having sized the population with AMPure before the ligation.

    The AMPure prior to the gel is likely to ensure that a majority of the ligase and free adapter (not to mention buffer/salt/DTT/ATP) are removed before sizing the sample. That much protein will almost certainly alter migration patterns...but by how much I can't say. It's an easy experiment to do though.

    Comment

    • ETHANol
      Senior Member
      • Feb 2010
      • 308

      #3
      Agree with ECO you need to do the AMPure purification before you run it on a gel. You are isolating small DNA fragments and you can't run the gel for very long or your gel slice will be too big. The salt front will likely run up against were you want to cut. If you want to cut corners, skip the gel purification.
      --------------
      Ethan

      Comment

      • Rocketknight
        Member
        • Sep 2011
        • 86

        #4
        Ah, good points, both of you. I'm afraid I'll have to keep the gel-cut size-selection step though. I'm using a Bioruptor rather than a Covaris and I'm doing exome enrichment followed by paired-end sequencing. I'm worried that if I allow big inserts through I'll end up getting a lot of reads that lie outside exons. Is that just paranoia?

        Comment

        • Heisman
          Senior Member
          • Dec 2010
          • 534

          #5
          Originally posted by Rocketknight View Post
          Ah, good points, both of you. I'm afraid I'll have to keep the gel-cut size-selection step though. I'm using a Bioruptor rather than a Covaris and I'm doing exome enrichment followed by paired-end sequencing. I'm worried that if I allow big inserts through I'll end up getting a lot of reads that lie outside exons. Is that just paranoia?
          If you sonicate on high for long enough I think you can get the Bioruptor size distribution down to a suitable size. May be worth a try.

          Comment

          • Rocketknight
            Member
            • Sep 2011
            • 86

            #6
            I can definitely get a peak in the size range I want, my main problem is that I feel the distribution will be broader than with Covaris, and so I'll get a lot of undesirably large inserts, even if the peak is in the right place.

            Comment

            • ECO
              --Site Admin--
              • Oct 2007
              • 1360

              #7
              Look into double sided AMPure cuts to avoid the gel....we do it all the time. They are described here and in the literature... I can't find the links now but that should get you started.

              Comment

              • Rocketknight
                Member
                • Sep 2011
                • 86

                #8
                Aha, awesome. I didn't realize bigger fragments outcompeted smaller ones when the beads were limited, I thought they just magically didn't bind fragments below a certain size or something. Is there a standard table for what bead:dna ratio I should use for each size cutoff, or should I determine my own?

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by Rocketknight View Post
                  Aha, awesome. I didn't realize bigger fragments outcompeted smaller ones when the beads were limited, I thought they just magically didn't bind fragments below a certain size or something. Is there a standard table for what bead:dna ratio I should use for each size cutoff, or should I determine my own?
                  The beads themselves are not the issue, are they? I understood that AmpureXP is a PEG cut with a fancy way to recover the precipitant, right?

                  Except I guess the beads themselves do have a maximum binding capacity. Anyone know what that is?

                  Oh, never mind, that is all covered here.
                  --
                  Phillip
                  Last edited by pmiguel; 02-02-2012, 08:46 AM.

                  Comment

                  • Rocketknight
                    Member
                    • Sep 2011
                    • 86

                    #10
                    Ahh, my bad again. It's the concentration of PEG in the Ampure XP buffer that governs what fragment size will precipitate. Sorry for the confusion.

                    Comment

                    • tbanks
                      Member
                      • Mar 2010
                      • 11

                      #11
                      This paper has some information on salt and PEG concentrations to size select using AMPure beads. They used a Bioanalyzer too.

                      Comment

                      • Rocketknight
                        Member
                        • Sep 2011
                        • 86

                        #12
                        Thanks all for the replies! I just started another batch of samples and got really nice size selection with Ampure. I sheared with a Bioruptor down to a peak size of about 260bp, then did the standard Illumina prep up until the first cleanup after end repair, then replaced their Ampure cleanup with the following:

                        1) Add 90ul Ampure XP to the 100ul end-repaired DNA.
                        2) Incubate 20 minutes
                        3) Separate beads and take supernatant
                        4) Add 20ul Ampure XP to the supernatant
                        5) Incubate 7 minutes
                        6) Separate the beads and clean them as normal (ethanol wash, dry, resuspend, etc)

                        This got me pretty much the perfect size range for 2x100bp exome sequencing. The only problem is I estimate I lost about 80-85% of my input DNA in the process. I have plenty of sample DNA to work with, but I'd prefer to only do the size-selection on DNA I've end-repaired. Does anyone know how well the Illumina end repair mix works if you use 4ug of sheared DNA at that step instead of 1ug?

                        Also, Bioanalyzer trace for the Ampure size-selected DNA:
                        Last edited by Rocketknight; 05-10-2012, 08:13 AM.

                        Comment

                        • Hamid
                          Senior Member
                          • Sep 2009
                          • 108

                          #13
                          Hi Rocketnight,

                          Did you by any chance do a Qubit analysis after you shearing to see if you are losing most of your DNA as single stranded DNA as a result of high energy over-shearing?
                          Comparing the Qubit results of your unsheared sample to the sheared sample will let you know if that is the cause of high loss of available sample for library preparation.

                          hamid

                          Comment

                          • Rocketknight
                            Member
                            • Sep 2011
                            • 86

                            #14
                            I checked that, yeah. I didn't lose much DNA in shearing - I think it just gets lost during double-Ampure selection. Do other people lose much less DNA with it?

                            Comment

                            • luc
                              Senior Member
                              • Dec 2010
                              • 469

                              #15
                              You might recover a bit more DNA if you reduce the first bead incubation time to 10 minutes. It is my impression that the size selection is anything but absolute - i.e. with longer incubation more DNA will keep on precipitating.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM
                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                09-28-2026, 10:25 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, Yesterday, 11:13 AM
                              0 responses
                              19 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, Yesterday, 10:48 AM
                              0 responses
                              18 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-15-2026, 10:52 AM
                              0 responses
                              36 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-14-2026, 05:39 AM
                              0 responses
                              42 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...